Re: HCS & in vivo imaging system

Posted by George McNamara on
URL: http://confocal-microscopy-list.275.s1.nabble.com/HCS-in-vivo-imaging-system-tp591433p591436.html

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Nishigandha,

Most papers imaging GFP or RFP in liver and other internal organs have been published by Robert M. Hoffman of AntiCancer Inc. http://www.anticancer.com/ (ex PubMed search: hoffman rm liver orthotopic). He typically uses skin flaps in his in vivo imaging models. See Bouvet M, Tsuji K, Yang M, Jiang P, Moossa AR, Hoffman RM. Cancer Res 66: 11293, for an example. He has published with the Olympus products and claims to have been involved in developing the FluorVivo ( http://www.anticancer.com/FluorVivoBrochure.pdf).

I have not yet seen the Xenogen system in action for in vivo fluorescence. If you want sensitivity by in vivo whole (intact) mouse imaging, go with Xenogen and firefly luciferase. If you want to be cutting edge (maybe bleeding edge), be the first to publish on a FLuc-mKate BRET model.

If you want to image single cells in lymph nodes in live mice at microscopy resolution, look at the Olympus IV100 with stick objective lens(es) (hopefully in multiphoton mode), or get yourself a multiphoton microscope. Not tumor papers, but see the reviews by Hauser, Shlomchik and Haberman 2007 Nat Rev Immunol 7: 499-504, and by Cahalan MD, Parker I. Annu Rev Immunol. 2008 Jan 2; [Epub] for starters. Multiphoton excitation will enable a lot deeper imaging into lymph node (or liver, if you can get the lens to it) than confocal - though mKate expressing or DiD, Cy5 or Cy5.5 labeled cells confocal imaging might not be too bad (long wavelength light scatters less).

If the Pan-a-See-Ya fits your requirements (i.e. imaging mice at macroscopic scale), and John Fox continues to decline to do a demo, ask if you can buy it on contingency - that is, if it works, you pay for it, if it does not, you ship it back (negotiate in advance who should be for shipping and whether a restocking fee is in order - after all, nce you've used it, it cannot be sold as new).

BD CARV HCS system (Pathway Bioimager) - I've seen the one in our building used for Fura-2 ratio calcium imaging. Looked ok. I just saw it in action for live pancreatic islet cell imaging, did not touch any knobs so can't say how sensitive it is compared to other systems (and did not see any focus changes, so cannot address how confocal it is). It was being run for sequential well timelapse assays, not HCS, so I cannot address how well it works for HCS.

best wishes,

George



At 04:29 AM 1/20/2008, you wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear George,

Thanks indeed for elaborate information.

Among high end in vivo imaging system, on papers
Xenogen did appear very good, especially for deep
seated bioluminescence, but I could not find reports
on imaging of GFP or RFP expressing metastatic cells.
I am interested on imaging metastasis of GFP or RFP
expressing cells in liver and lymphnodes. What is your
opinion and experience reagrding this application with
Xenogen system?

Among lower end systems, FluorVivo of INDECO
BioSystems and Pan-a-See-Ya of Lightools looked good.
But, I could not get demo of Pan-a See-Ya from John
Fox. Do you know any users who have sucessfuly used
any of these for imaging metastasis in lymphnodes and
liver? 

Do you have any infromation on performance HCS system
of BD that has in built Nipkow disc based confocal
system?

Once again thanks a lot for the information.
With best regards,
Nishigandha Naik

--- George McNamara <[hidden email]> wrote:

> Search the CONFOCAL archive at
>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Nishigandha,
>
> As always - get a demo! Really more a question of
> budget and throughput.
>
> For in vivo, the Caliper/Xenogen IVIS Spectrum looks
> very good.
>
http://www.caliperls.com/products/optical-imaging/ivis-spectrum.htm
>
> My colleagues at UM are upgrading our IVIS 200 to
> the Spectrum in a
> month. Advantage of the IVIS systems is that using
> cryogenic
> back-illuminated CCD's enables luciferase
> bioluminescence imaging
> (BLI) in vivo (and ex vivo - extremely valuable to
> take out the
> organs at the end and re-image without the rest of
> the animal in the
> way. also, can image multi-well plates, so can test
> the performance
> of cell lines before putting them in the mouse).
> A huge advantage of the Xenogen systems is that they
> have made the
> effort to provide quantitative data for
> bioluminescence. They use
> photons/second/cm^2/steradian. This enables every
> paper published
> with a Xenogen system to be compared to each other
> (of course, the
> depth of the cells, expression level, and amount of
> luciferin at the
> cells are additional variables). If you get a
> non-Xenogen BLI
> instrument, make sure the vendor validates the same
> units. Xenogen
> now does something similar for fluorescence
> tomography (a trickier issue).
>
> CRi has been in the in vivo fluorescence field for
> several years.
> Check out their
> http://www.cri-inc.com/products/maestro2.asp
>
> If your budget is on the low end, arrange a demo
> from John Fox of
> Lightools ( http://www.lightools.com/tutorial.htm). I
> was especially
> impressed by a demo of his a couple of years ago,
> that included the
> Pan-a-See-Ya
> ( http://www.lightools.com/LRTPDFS/panaseeypdf.pdf),
> a
> simple RGB color camera, and mouse in hand without
> anesthesia (you
> need an isofluorene rig for the other systems).
>
> Reagents, cells and animals for in vivo imaging:
>
> Marker Gene Technology and Promega both sell
> gal-luciferin, a LacZ
> substrate whose product is a firefly luciferase
> substrate. Sam
> Gambhir has lots of tribrid vectors of
> FP-luciferase-TK (TK = enzyme
> for use with PET reporter and therapy).
> Xenogen also has several transgenic mice and various
> cell lines
> expressing luciferase. GFP (and maybe by now RFP)
> and firefly
> luciferase transgenic mice are available from
> Jackson Labs
> (www.jax.org). Lots of plasmids are available from
> www.addgene.com
> Which FP works best is tissue dependent - see
> http://www.jhc.org/cgi/content/abstract/55/9/931 for
> one study.
> You can also demo by injecting known volumes of
> fluorescent beads
> (i.e. in matrigel) under the skin, though you should
> expect most of
> the research to be of deeper sites. Ralph
> Weissleder's company, Visen
> Medical, sells a variety of fluorescence reagents
> for mouse in vivo
> imaging (AngioSense 680, etc). Fluorescence
> angiography with
> indocyanine green (ICG) or fluorescein
> ( http://msp.rmit.edu.au/Article_02/03d.html) has
> been clinically
> available for years, and is a good way to start with
> mice.
>
>
>
> HCS - Cellomics (Thermo Fisher) founded the field,
> worth demoing.
> They had a nice article on their web site on how
> adding an Apotome
> (Zeiss optical sectioning device) to their
> instrument improved
> Z'-scores, so confocal imaging should help.  Drug
> Discovery Today,
> ASSAY and Drug Development Technologies, Nature
> Reviews Drug
> Discovery, and a bunch of other journals cover this
> field. Start with
> the following books:
> Methods Enzymol. 2006;414
> Methods Mol Biol. 2007;356
> Then, search PubMed for   high content screen*
>
>
> George
>
>
> At 12:56 AM 1/12/2008, you wrote:
> >Search the CONFOCAL archive at
>
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> >
> >Dear All,
> >
> >I am planning to procure High Content Screening
> system
> >(HCS). I would like to know experience of users as
> I
> >could not find many published reports on different
> >systems. I am also debating on utility of confocal
> >(Nipkow disk based) based HCS versus without
> confocal
> >HCS. I will be thankful if anybody could share
> their
> >experience or opine on the various available makes
> in
> >the market.
> >
> >I am also looking for in vivo imaging system to
> >procure images from small animals such as mouse,
> rats
> >to study tumor formation, metastasis and
> angeogenesis
> >of GFP or RFP expressing. I request to share your
> >experience to enable me to choose the proper system
> >for these applications.
> >
> >With thanks and best wishes for the year,
> >
> >Nishigandha Naik
> >
> >
> >
> >
> >
>
>____________________________________________________________________________________
> >Looking for last minute shopping deals?
> >Find them fast with Yahoo!
> >Search.
>
http://tools.search.yahoo.com/newsearch/category.php?category=shopping
>
>
>
>
>
>
> George McNamara, Ph.D.
> University of Miami, Miller School of Medicine
> Image Core
> Miami, FL 33010
> [hidden email]
> [hidden email]
> 305-243-8436 office
> http://home.earthlink.net/~pubspectra/
> http://home.earthlink.net/~geomcnamara/
>
http://www.sylvester.org/health_pro/shared_resources/index.asp
> (see
> Analytical Imaging Core Facility)
>



      ____________________________________________________________________________________
Never miss a thing.  Make Yahoo your home page.
http://www.yahoo.com/r/hs




 

George McNamara, Ph.D.
University of Miami, Miller School of Medicine
Image Core
Miami, FL 33010
[hidden email]
[hidden email]
305-243-8436 office
http://home.earthlink.net/~pubspectra/
http://home.earthlink.net/~geomcnamara/
http://www.sylvester.org/health_pro/shared_resources/index.asp (see Analytical Imaging Core Facility)