hours. We could see confocal mitosis and all the nice biology of GFP
phototoxicity. More over Eric is a very nice person to collaborate with.
> Search the CONFOCAL archive at
>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>
> Hi David,
>
> the system measures the amount of fluorescence signal per pixel and
> controls the AOTF for this pixel based on the measured value. So if
> there is no light coming to the detector, laser will be restricted from
> illuminating the sample. Or, if the detector gets close to saturation,
> laser will be restricted as well. This should reduce photobleaching and
> phototoxicity also in out-of-focus areas. That's at least how I
> understand this.
>
> "The CLEM control unit is an optional add-on system for the C1. The
> CLEM unit calculates the integrated detected signal and exposure
> time for each pixel in real time, performs AOM high speed shutter
> control and high speed operation processing based on the amount of
> acquired fluorescence signal, PMT gain and then outputs the
> resulting fluorescence signal to the C1 controller’s line grabber
> electronics circuit."
>
> source:
http://www.nikoninstruments.eu/news.php?n=550>
> I expect the quantification (measuring intensities) to be difficult
> afterwards, since one needs to apply a correction. However, it's one of
> the main things that makes the Nikon system interesting.
>
> Michael
>
>
> David Knecht wrote:
>> Search the CONFOCAL archive at
>>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>>> The ability of using this system along with Controlled Light Exposure
>>> Microscopy (CLEM) makes this system for long time live cell imaging and
>>> confocal analysis. CLEM automatically monitors and varying the laser
>>> illumination during time-lapse studies to reduce the risk of laser induced
>>> bleaching, biochemical inconsistency and cell death.
>>
>> Can you clarify what this means? I don't see how you can monitor
>> the imaging and vary the laser in a way that can avoid cell
>> toxicity. Typically, you don't want the laser intensity to vary.
>> Do you have any quantitative data on cellular toxicity during
>> imaging in comparison to a spinning disk system?
>>
>> Dr. David Knecht Department of Molecular and Cell Biology
>> Co-head Flow Cytometry and Confocal Microscopy Facility
>> U-3125
>> 91 N. Eagleville Rd.
>> University of Connecticut
>> Storrs, CT 06269
>> 860-486-2200
>> 860-486-4331 (fax)
Dep. Moleculaire Biotechnologie