Hi Stan,
On Jan 14, 2011, at 7:02 AM, CONFOCALMICROSCOPY automatic digest system wrote:
>
> Date: Thu, 13 Jan 2011 15:28:40 -0600
> From: Stanislav Vitha <
[hidden email]>
> Subject: Re: Objective lens chromatic aberration - Color shift correction for colocalization analysis - was Glycerol Objectives - experience with
>
> Hi Dan,
> The translational correction is very useful, but what if the shift varie=
> s
> across the field?
then the assumption of the shift being space invariant is clearly false,
and you need a more clever way to correct it, that IS spatially variant... read on.
> Do you know if there is a plugin or a script to perform affine transform =
> in
> 3D?
Oh yes!
> The goal would be to "warp" one channel in XYZ to match the second ch=
> annel.
There are options here...
In Fiji (i just ImageJ) there is Ignacio's nice bUnwarpJ
http://pacific.mpi-cbg.de/wiki/index.php/BUnwarpJpoints to
http://biocomp.cnb.csic.es/~iarganda/bUnwarpJ/index.htmldont think it can also warp z though...
> I have imaged 100 nm multi-color beads on our confocal (FV1000 with a
> hand-picked 100x/1.4 Plan Super Apo lens), and besides a moderate z-shift=
>
> between blue and other channels (as expected), I have a lateral chromatic=
>
> aberration between the "DAPI" channel (405 nm ex, 430-460 em) and the lon=
> ger
> wavelength channels (green, orange, red fluorescence). So for instance, t=
> he
> colors are colocalized in top left corner of the image, but are several
> pixels off in XY plane in bottom right of the image. The UnwarpJ plugin f=
> or
> ImageJ works well to correct this in individual XY images,
not the same as bUnwarpJ, but similar....
> but 3D correct=
> ion
> may be needed, since the plan correction for the blue channel (405 nm las=
> er
> illumination) is not as good as for the rest of the spectrum.=20=20=20
thats often the case,
as well as the UC laser coming in through a different collimator, so its often 0.5 microns or so off the vis lines.
cheers
Dan
>
> =20
>
> Stan Vitha
> Microscopy and Imaging Center
> Texas A&M University
Dr. Daniel James White BSc. (Hons.) PhD
Senior Microscopist / Image Visualisation, Processing and Analysis
Light Microscopy and Image Processing Facilities
Max Planck Institute of Molecular Cell Biology and Genetics
Pfotenhauerstrasse 108
01307 DRESDEN
Germany
+49 (0)15114966933 (German Mobile)
+49 (0)351 210 2627 (Work phone at MPI-CBG)
+49 (0)351 210 1078 (Fax MPI-CBG LMF)
http://www.bioimagexd.net BioImageXD
http://pacific.mpi-cbg.de Fiji - is just ImageJ (Batteries Included)
http://www.chalkie.org.uk Dan's Homepages
https://ifn.mpi-cbg.de Dresden Imaging Facility Network
dan (at) chalkie.org.uk
( white (at) mpi-cbg.de )