Posted by
James Pawley on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Re-Objective-lens-chromatic-aberration-Color-shift-correction-for-colocalization-analysis-was-Glycerh-tp5919806p5922644.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Hi all,
Chromatic aberration is a measure of how the focal length (f) of a
lens changes with wavelength. As the magnification of any optical
system depends on the focal length of the lenses involved, of course
this means that, to the extent that chromatic aberration is present,
the magnification will change with wavelength. In simple optical
systems it is visible as the pronounced red or blue fringes around
the edges of images of high-contrast features. This is called lateral
chromatic aberration and, if the optical system is properly aligned,
it "should" be essentially invisible in the center of the field and
get larger as you approach the edges. (The finding that it lines up
in one corner but not elsewhere makes me think that there may be an
alignment issue.)
As mentioned before all refractive lenses have some chromatic
aberration. (Reflective lenses do not). A good plan-apo, may be
corrected for 4 wavelengths but the wavelengths chosen may vary from
lens to lens. Between these wavelengths there will show
wavelength-dependent shifts in focal length. The manufacturer
endeavors to keep these shifts less than the depth of field of the
objective so that they are not annoying to the naked eye when viewing
multicolored specimens. In practice you will only see the effects
when viewing planar objects such as patterns in thin metal films that
have been evaporated onto the glass surface.
Alan Boyde was the first to point out that these variations in f were
found in even the most expensive objectives, using a tandem-disk
scanner with Hg illumination (which has a number of intensity peaks
in the visible) and making backscattered/reflected light images of
planar features (surface of the slide). In fact he used this
"problem" to make the first truly live-time 3D confocal microscope:
Lenses were designed to have pronounced linear chromatic aberration.
Consequently, different planes in the specimen satisfied the confocal
conditions at different wavelengths. He then used prisms or
diffraction elements in the oculars that were oriented so that, say,
the red image was slightly displaced to the inside (towards the other
eye) of the blue. Looking through the oculars, the viewer then saw a
stereo image, obtained at full NA that was also color-coded for
topographic height. Because confocal images could be obtained at
several focal depths simultaneously, it was no longer necessary to
move the specimen in z to obtain a focal series, making live-stereo
images possible. These livetime stereo images could even be recorded
on tape and shown to groups using special glasses
From this you can see that any attempt to correct for CA must be able
to respond to changes in magnification with wavelength as well as the
vertical displacements between the z-location of the on-axis peak
confocal signal vs wavelength. As noted above, there may also be
misalignment issues.
Anyone still willing to try to undertake this task should spend a lot
of time studying the work of Cremer and Cremer in the 1980s and
1990s. These scientists demonstrated one of the first effective
systems of super-resolution widefield and confocal imaging as applied
to FISH (fluorescent in-sit hybridization). They labelled one gene on
a chromosome red, and another nearby one green. They then recorded
the green spot and found its centroid. DItto, the red spot. Because
these two spots were separated by wavelength, their images did not
overlap in the recording system and therefore the distance between
them could be measured with an accuracy that depended on the ability
to determine the position of the centroid of each spot, rather than
on the Abbe resolution of the optics. However, as the position of the
centroid did depend on the CA of the optical system, they had to make
very careful measurements of this parameter in order to apply the
proper correction factors.
Hope to see some of you in Vancouver next June!
Cheers,
Jim Pawley
***************************************************************************
Prof. James B. Pawley, Ph.
608-238-3953
21. N. Prospect Ave. Madison, WI 53726 USA
[hidden email]
3D Microscopy of Living Cells Course, June 11-23, 2011, UBC, Vancouver Canada
Info:
http://www.3dcourse.ubc.ca/ Applications due by March 15, 2011
"If it ain't diffraction, it must be statistics." Anon.
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>*****
>
>Hi Dan,
>The translational correction is very useful, but what if the shift varies
>across the field?
>Do you know if there is a plugin or a script to perform affine transform in
>3D? The goal would be to "warp" one channel in XYZ to match the
>second channel.
>I have imaged 100 nm multi-color beads on our confocal (FV1000 with a
>hand-picked 100x/1.4 Plan Super Apo lens), and besides a moderate z-shift
>between blue and other channels (as expected), I have a lateral chromatic
>aberration between the "DAPI" channel (405 nm ex, 430-460 em) and the longer
>wavelength channels (green, orange, red fluorescence). So for instance, the
>colors are colocalized in top left corner of the image, but are several
>pixels off in XY plane in bottom right of the image. The UnwarpJ plugin for
>ImageJ works well to correct this in individual XY images, but 3D correction
>may be needed, since the plan correction for the blue channel (405 nm laser
>illumination) is not as good as for the rest of the spectrum.
>
>
>
>Stan Vitha
>Microscopy and Imaging Center
>Texas A&M University
>
>
>On Tue, 11 Jan 2011 11:23:02 +0100, Daniel James White
><
[hidden email]> wrote:
>
>>*****
>>To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>*****
>>
>>Hi Martin,
>>
>>On Jan 11, 2011, at 7:01 AM, CONFOCALMICROSCOPY automatic digest
>>system wrote:
>>
>>> Date: Mon, 10 Jan 2011 17:18:29 -0600
>>> From: Martin Wessendorf <
[hidden email]>
>>> Subject: Re: Glycerol Objectives - experience with
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>> *****
>>>
>>> On 1/10/2011 4:32 PM, Rosemary White wrote:
>>>
>>>> I've found that the red and blue emission aren't quite lined up
>>>>vertically,
>>>> at least with our "blue" 63x objective, so have to cut the top one or two
>>>> slices off a blue vertical series (depending on depth of slice), and the
>>>> bottom slices off a red series and reassemble to get the
>>>>emissions aligned -
>>>> i.e. from the same depth in the tissue. The objectives do vary a bit,
>>>> perhaps ours isn't as well corrected as some.
>>>
>>> Can anyone explain where the difficulty arises in correcting for axial
>>> chromatic aberration? I've consistently seen serious axial chromatic
>>> aberration in good quality oil objectives from one very reputable
>>> manufacturer, between red, green and far-red (1 um off in the z-axis
>>> between green and red, and between red and far-red; 2 um between green
>>> and far-red). I have always heard that axial chromatic aberration was
>>> easy to correct for and would've thought that a solution for 3-color
>>> correction would've been found 100 years ago. However, I don't know
>>> enough optics to understand the subtleties. Are there trade-offs to
>>> trying to obtain good correction, besides the cost in scattering of
>>> adding additional lens elements?
>>
>>Sadly, not all objective lenses were made equal.
>>
>>In our hands on point scanning confocals, even when you align the
>pinhole(s) as best you can
>>there can still be a micron of misalignment in between "green", "red" or
>"far red",
>>with DAPI etc. often being way off, as it comes through a different optic
>fiber and collimator anyway (eg on an LSM 510)
>>
>>I have tested different lenses of the same spec from the same manufacturer,
>>and found that they all have their own personality.
>>Good correction is very possible, but "good" is a relative thing.
> >
>>I send back ones that are too "bad", and keep the "good" ones.
>>I take z stacks of 1 micron multi colour beads to measure the
>>remaining error
>>(dont need sub resolution bead images for this measurement)
>>
>>There will always be a significant error, even in the very best
>>"Super-dooper Mega Extra Apo-Chromat" lens you can get from any
>manufacturer for less than 100 k dollars.
>>They simply can not be made perfect at a reasonable cost (as explained to
>me by a Zeiss lens guru)
>>
>>If you want to do precise colocalization studies at the highest optical
>resolution (everyone does, even if they don't immediately realize it)
>>then you simply MUST measure the error, then correct/shift the images (in
>3D) before colocalization/correlation analysis.
>>
>>This can be done in 3D in ImageJ/ Fiji
>>using nice interpolation methods from Erik Meijering, for sub pixel shifts
>- TJ-Translate :
>>
http://pacific.mpi-cbg.de/wiki/index.php/TransformJ>>leading to
>>
http://imagescience.org/meijering/software/transformj/translate.html>>
>>or in the great colour shift corrector of Huygens Professional (no
>commercial interest - just a happy customer)
>>
http://www.svi.nl/ChromaticShiftCorrector>>
>>or roughly in the Zeiss AIM 510 software in xy only (whole pixel shifts),
>>
>>The effect on the 2 channel scatterplot / 2D histogram / fluorogram is very
>significant.
>>A ugly poorly correlated cloud turns into tight correlated populations of
>pixels,
>>and the coloc. coefficients jump much higher.
>>
>>more info here:
>>
http://ifn.mpi-cbg.de/wiki/ifn/index.php/Chromatic_aberration_measurement_and_correction>>
>>cheers
>>
>>Dan
>>
>>
>>
>>>
>>> Thanks--
>>>
>>> Martin
>>> --
>>> Martin Wessendorf, Ph.D. office: (612) 626-0145
>>> Assoc Prof, Dept Neuroscience lab: (612) 624-2991
>>> University of Minnesota Preferred FAX: (612) 624-8118
>>> 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009
>>> Minneapolis, MN 55455 e-mail:
[hidden email]
>>
>>Dr. Daniel James White BSc. (Hons.) PhD
>>Senior Microscopist / Image Visualisation, Processing and Analysis
>>Light Microscopy and Image Processing Facilities
>>Max Planck Institute of Molecular Cell Biology and Genetics
>>Pfotenhauerstrasse 108
>>01307 DRESDEN
>>Germany
>>
>>+49 (0)15114966933 (German Mobile)
>>+49 (0)351 210 2627 (Work phone at MPI-CBG)
>>+49 (0)351 210 1078 (Fax MPI-CBG LMF)
>>
>>
http://www.bioimagexd.net BioImageXD
>>
http://pacific.mpi-cbg.de Fiji - is just ImageJ
>>(Batteries Included)
>>
http://www.chalkie.org.uk Dan's Homepages
>>
https://ifn.mpi-cbg.de Dresden Imaging
>>Facility Network
>>dan (at) chalkie.org.uk
>>( white (at) mpi-cbg.de )
--
***************************************************************************
Prof. James B. Pawley, Ph.
608-238-3953
21. N. Prospect Ave. Madison, WI 53726 USA
[hidden email]
3D Microscopy of Living Cells Course, June 11-23, 2011, UBC, Vancouver Canada
Info:
http://www.3dcourse.ubc.ca/ Applications due by March 15, 2011
"If it ain't diffraction, it must be statistics." Anon.