Re: 2P standard

Posted by mcammer on
URL: http://confocal-microscopy-list.275.s1.nabble.com/2P-standard-tp5923717p5924007.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Fluorescent plastic slides are not stabile.  Depending on their exposure to light they bleach or change color.  Using a pulsed laser is a good way to make little burn or other chemical reaction spots in them.  Here are two old web pages with some of the problems (and benefits!) explained:  
http://www.einstein.yu.edu/aif/instructions/fluor-ref-slides/01.htm 
http://www.einstein.yu.edu/aif/instructions/zeiss/liveduo/Zaxis_bleachingtest/index.htm
_______________________________________
Michael Cammer, Assistant Research Scientist
Skirball Institute of Biomolecular Medicine
Lab: (212) 263-3208  Cell: (914) 309-3270

________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Mark Cannell [[hidden email]]
Sent: Friday, January 14, 2011 7:39 PM
To: [hidden email]
Subject: Re: 2P standard

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

How about Fluorescent plastic slides -I got some free from Chroma.

Cheers


On 15/01/2011, at 12:34 PM, Boswell, Carl A - (cboswell) wrote:

> Hi all,
> I'm looking for some trustworthy, permanent standard(s) to use in a
> MP system to monitor the condition of detectors and the system in
> general.   The goal is a fluorescent source that is immutable.  This
> could done either by having unrestricted replacement of bleached
> label with unbleached label (i.e. a solution), or a uniform and very
> large solid, so that it is unlikely that any one tiny volume will be
> imaged repeatedly.   We've tried several iterations of an aqueous
> solution of label, but sealing the prep does not prevent eventual
> evaporation.  Plus, there seems to be some modification of FITC, for
> example,  in solution over time (a month) regardless of exposure to
> light.  I'm not sure I can trust the consistency of colored plastic
> slides, since they are not manufactured with this role in mind.  One
> possibility is uranyl glass slides, but Corning says they don't make
> these any more.  Does anyone know a source?  We're also considering
> quantum dots in something like immersion oil , in a sealed chamber
> but I don't have any experience with these labels in a hydrophobic
> environment.
>
> Thanks for your insights.
> C
>
>
> Carl A. Boswell, Ph.D.
> Molecular and Cellular Biology
> Univ. of Arizona
> 520-954-7053
> FAX 520-621-3709
>
>

------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================