Dear Monique,
We need to use confocal and collect reflected laser. One will need to setup a light path that allows passage of reflected laser to a detector which has pin-hole in front - which should be possible on most of the microscopes as the dichroic never completely gets rid of excitation laser. At air-glass and glass-air interfaces, laser will be reflected much more strongly than within coverslip. This residual laser from dichroic should be detectable if the emission filter is removed from the path. The confocal pin-hole will provide strong depth descrimination and allow measurement of coverslip to submicron accuracy.
The correction factor that I mentioned assumes normal incidence of laser on coverslip, so the use 'beam-parking' or 'point-scanning' mode at the center of field of view is required.
Regards
Shalin
On Mon, Feb 25, 2008 at 11:12 PM, Vasseur Monique <
[hidden email]> wrote:
How to do you that? See it?
What is the microscope setup?
Monique Vasseur
Microscopie et imagerie
Département de biochimie
Université de Montréal
C.P. 6128, succursale Centre-ville
Montréal QC H3C
3J7 Canada
tél. (514) 343-6111 poste 5148
De :
Confocal Microscopy List [mailto:[hidden email]] De la part de Rietdorf, Jens
Envoyé : 24 février 2008
15:30
À : [hidden email]
Objet : Re: Coverslip
Dear Sarah,
you could use reflections from the
coverglass which are characteristic for upper and lower edge of the cover glass
and easy to spot.
regards,
jens
---
Dr.
Jens Rietdorf[hidden email]
Head Microscopy
Novartis Research Foundation
Friedrich-Miescher-Institute,
wro1066.2.16 Maulbeerstr.66, CH-4058 Basel, Switzerland
phone +41(61)69-75172 mobil +41 798284737
Email:rietdorf(at)fmi.ch
From:
Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sarah Kefayati
Sent: Freitag, 22. Februar 2008
21:42
To: [hidden email]
Subject: Coverslip
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
I need to know how I can measure the depth from the cover slip,is there
any marked or fluorescent coverslip that I can use for this purpose?
Thanks for any suggestion
--
~~~~~~~~~~~~~~~~~~~~~~~~~
Shalin Mehta
mobile: +65-90694182
blog:
shalin.wordpress.com~~~~~~~~~~~~~~~~~~~~~~~~~~
Bioimaging Lab, Block-E3A, #7-10
Div of Bioengineering, NUS Singapore 117574
website:
http://www.bioeng.nus.edu.sg/optbioimaging/colin/index.html
Liver Cancer Functional Genomics Lab, #6-05
National Cancer Centre, Singapore 169610
~~~~~~~~~~~~~~~~~~~~~~~~~~~
What is light?
''Light behaves like Voltaire, the French genius, 250 years ago. He was born as a conservative Catholic, he lived as a liberal Protestant, and he returned to Catholicism when he faced the end of his life. Light is born as photons. It travels through space as a wave, and it dies finally as a photon.'' - Einstein as quoted in Lohmann's paper (JOSA-A, Vol. 17, No.10, pp.1755-1762)