refractive index than it's surroundings. That being said, would it be
good one. While I don't believe it to be an issue here (as the
> Search the CONFOCAL archive at
>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>
> Since some parts of the peroxisomal lumen are comprised of a
> "crystalline matrix" of protein, would this affect the refractive
> index and thus create spherical aberration in the PSF?
>
> Michael J. Schell, Ph.D., CIV, USUHS
> Assist. Professor
> Dept. of Pharmacology
> Uniformed Services University
> 4301 Jones Bridge Rd.
> Bethesda, MD 20814-3220
> tel: (301) 295-3249
>
[hidden email]
>>>> Glen MacDonald <
[hidden email]> 02/22/08 5:47 PM >>>
> Search the CONFOCAL archive at
>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>
> Hi Fred,
> Do you know for certain that the chosen peroxisome has a homogenous
> distribution of fluoropore? Looking at isolated peroxisomes might
> give a hint and provide a nice control as to whether the distortion
> of the PSF in comparison to plain beads is due to the optical
> properties of the sample. And, as you estimate, presenting a more
> realistic PSF. I'm assuming you are looking at fixed yeasties, or
> John's concern regarding motion would be very valid.
>
> Regards,
> glen
> Glen MacDonald
> Core for Communication Research
> Virginia Merrill Bloedel Hearing Research Center
> Box 357923
> University of Washington
> Seattle, WA 98195-7923 USA
> (206) 616-4156
>
[hidden email]
>
> ************************************************************************
> ******
> The box said "Requires WindowsXP or better", so I bought a Macintosh.
> ************************************************************************
> ******
>
>
> On Feb 22, 2008, at 1:54 PM, Fred Mast wrote:
>
>> Search the CONFOCAL archive at
>>
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal>>
>> Hello all,
>>
>> I would like some input into an idea I have for obtaining a PSF. I
>> study peroxisomes using yeast as a model system for understanding
>> how they are created. In yeast, peroxisomes have a well
>> characterized morphology, being spherical organelles with a
>> diameter of 100 to 200nm. Other than this size variability I think
>> they are excellent candidates for obtaining a PSF as they can be
>> easily, fluorescently labelled (by targeting fluorescent protein
>> chimeras to their matrix), are similar in size to what is typically
>> used to obtain PSF's, and are "embedded" in the sample. I do a lot
>> of live cell imaging, using a LSM510 Meta and am always looking for
>> ways to improve my system. As a result most of my images are fairly
>> noisy and I rely on deconvolution to remove the noise, and improve
>> contrast and resolution. My initial attempts at using peroxisomes
>> for this purpose have provided me with a PSF that is slightly
>> different from what I obtain with fluorescent beads (the peroxisome
>> derived PSF is less symmetrical) and provides, in my estimation, a
>> more realistic result. Your thoughts and concerns on this idea
>> would be most welcome.
>>
>> Fred
>>
>> Fred D. Mast
>> Department of Cell Biology
>> Medical Sciences Building Room 5-14
>> University of Alberta
>> Edmonton, Alberta, T6G 2H7
>> Canada
>>
>> Tel: 1-780-492-7407
>>
[hidden email]
>
Fred D. Mast