Re: Detector question -- APD

Posted by James Pawley on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Alexa-Fluor-488-solubility-problem-tp592502p592512.html

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Dear Guy,
>
>Just to make it more precise: You can do APD-imaging also with a
>Zeiss-confocal if you have the ConfoCor3 FCS/FCCS setup. IT works
>very nicely and gives you indeed a huge signal increase.
>
>Cheers    Gabor
>
>--
>Gabor Csucs Light Microscopy Centre, ETH Zurich
>Schafmattstrasse 18, HPM F16 CH-8093, Zurich, Switzerland
>
>Web: www.lmc.ethz.ch
>Phone: +41 44 633 6221
>Fax: +41 44 632 1298
>e-mail: [hidden email]

I agree that ConFoCOr seems an appropriate use: the signal level is
very low indeed. And the contrast is basically the presence or
absence of a single fluorophor in the spot.

The STED also is noted for low signal levels as most of the
excitations are "entrained" by stimulated emission to go away from
the detector.  On the other hand, even with a 30MHz counter you can
conceivably count up to about 20 pulses in a microsecond, before the
pileup losses become too obvious to the eye, while a faster circuit
or a longer pixel-dwell time allow you to count proportionally more.

Cheers,

Jim P.
--
               **********************************************
Prof. James B. Pawley,                          Ph.  608-263-3147
Room 223, Zoology Research Building,              
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/             Applications still being accepted
               "If it ain't diffraction, it must be statistics." Anon.