Dave,

Depends on what you mean by easy. To narrow the problem a bit, the important issue is how much UV reaches the sample. Any other filter is more or less irrelevant with respects to cell viability, unless there is so much back scattered UV that extra background noise must be averaged out with extra exposures.

Ignoring the latter, one might deal with the former by using a test object that is excited by UV. There are quite a few options including UV sensitive inorganic phosphors that can be purchased as fine powders or disks. The latter compounds can have luminescent lifetimes that range from as short as 40 nsec. to milliseconds. The latter often use a lanthanide such as europium (red) or terbium (blue, green, red) with sharp emission lines. Some have broad emissions such as P31, a ZnS based phosphor. Another possibility would be some organic fluorophores including AMCA or well saturated nuclei stained with DAPI or Hoechst (say 3 ug dye /ml of cells).

A slide made with a phosphorescent disk will last for years. As I recall, some EM suppliers sell phosphorescent disks and related materials for use in cathodoluminescent detectors. I have used or made all of the above options in one form or another. A further option is to use a fiber optic couple micro spectrofluorometer. I like the USB ported version from Ocean Optics, Fl. (no financial interest). They can provide a cable that is capped with a collection lens that carries the excitation light from the microscope objective to the spectrometer which uses a grating and linear CCD. As I recall, there are a few grating options one of which provides sensitivity down near 220 nm on up to 700 nm.

Anyway, some of these approaches are crude but they will tell you if there is significant UV bleed through.

Mario


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have used the CoolLED system, and it is not operating in strobe mode.  It is continuous illumination.  The UV question is interesting.  Is there an easy way to test if there is UV leaking through the epi filter set with the mercury burner?  Dave

On Mar 12, 2008, at 4:27 PM, Stanislav Vitha wrote:

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Regarding the difference between LED illumination and Hg lamp -
Perhaps if the illumination is not continuous but is set up for strobe
operation of the LED, this could allow dark state relaxation an prolong
the life of the fluorophore (I am not sure the same would apply to the
life of the cell). - The time between light pulses should be more than one
microscecond.
reference:   
Gerald Donnert, Christian Eggeling & Stefan W Hell: Major signal increase
in fluorescence microscopy through dark-state relaxation. Nature Methods -
4, 81 - 86 (2007)


Stan Vitha

Dr. David Knecht  
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)


-- 
________________________________________________________________________________
Mario M. Moronne, Ph.D.
cell (510) 367-8497

[hidden email]
[hidden email]
[hidden email]