I have observed as well that people in
general cannot see out into the far red wavelengths. Sometimes if the sample
is very bright, you can see it, but in most cases not. The far-red cubes do
work very well however for most ccd cameras. Be aware that many cameras have a
blue filter installed to reduce IR reaching the ccd. This might have to be
removed (consult manufacturer) in order to image in the far-red.
Cheers,
Owen
Owen M. Schwartz PhD
Head, Biological Imaging Facility
NIAID-NIH
Bldg 4 room B2-30
Ph 301-402-5883
Fax 301-480-2882
From: Kenneth Bowitz
Larsen [mailto:[hidden email]]
Sent: Friday, March 07, 2008 6:43
AM
To: [hidden email]
Subject: Far-red filter set
Hi,
we're
considering buying a Y5 filter cube from Leica to see the far-red Alexas, DRAQ5
etc. but then someone mentioned to me that many (most?) people are unable to
see far-red light. Does anyone know if this is true, or have experience with
this (or a similar) filter cube? It has a BP 700/75 suppression filter.
Kenneth