http://confocal-microscopy-list.275.s1.nabble.com/Re-Objective-lens-chromatic-aberration-Color-shift-correction-for-colocalization-analysis-was-Glycerh-tp5919806p5929596.html
point. Yes, when imaging a point object that
planes (for different colors) in the specimen.
symmetrical disk). Likewise, a point scatterer
also be in focus when it reaches the disk plane.
codes for specimen topography. In this case you
be corrected in the ocular. And that using such
effects in early BSL imaging using disk-scaners.
single color. Although this produced stunning 3D
were as good as Guy says. However they also had
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>*****
>
>Alan Boyde's TSM work, while fun and
>interesting, does not accurately show up the CA
>of the lens when used as it is intended. Abbe's
>original apochromat design corrected axial
>chromatic aberration at 3 wavelengths (and SA at
>2). But it left a lot of lateral chromatic
>aberration (chromatic difference of
>magnification). This was corrected by matching
>eyepieces (compensating eyepieces) so that one
>saw a beautifully corrected image. (With normal
>eyepieces the result was woeful).
>
>
>
>This principle has been followed by most
>manufacturers ever since, and to avoid the need
>to swap eyepieces achromat and fluorite lenses
>were designed to require the same correction
>(even though they didn't need to be). When the
>change was made to infinite tube length some
>(not all) manufacturers switched the correction
>to the tube lens.
>
>
>
>What has this to do with Boyde's TSM? In a
>PetrĂ n type TSM, the spinning disk lies between
>the eyepieces and the objective, so the pinhole
>image is not corrected. This means that
>in-focus spots at the red and blue ends of the
>spectrum will miss their respective pinholes.
>The resulting colour patterns can be very pretty
>but are quite complex to interpret. They do
>not tell us what the lens will do when used
>correctly.
>
>
>
>When I learnt the principles of microscopy (from
>the legendary JR Baker, 45 years ago) I was
>introduced to aberration diagrams, which show
>the exact focal length excursion with
>wavelength. Back then these were just a useful
>figure of merit for lenses. Now, with advanced
>3-D microscopy, they are essential. But it is
>very hard to get them out of manufacturers. I
>have once - only - seen a manufacturer show one
>of these curves (and a very impressive one it
>was) but that was only in a slide - I've never
>got one on paper. We really need to pressure
>manufacturers into giving us this information.
>
>
>
>
>Guy
>
>
>
>Optical Imaging Techniques in Cell Biology
>
>by Guy Cox CRC Press / Taylor & Francis
>
>
http://www.guycox.com/optical.htm <
http://www.guycox.com/optical.htm>
>
>______________________________________________
>
>Associate Professor Guy Cox, MA, DPhil(Oxon)
>
>Australian Centre for Microscopy & Microanalysis,
>
>Madsen Building F09, University of Sydney, NSW 2006
>
>
>
>Phone +61 2 9351 3176 Fax +61 2 9351 7682
>
> Mobile 0413 281 861
>
>______________________________________________
>
>
http://www.guycox.net <
http://www.guycox.net>
>
>
>
>
>
>From: Confocal Microscopy List
>[mailto:
[hidden email]] On
>Behalf Of James Pawley
>Sent: Saturday, 15 January 2011 4:30 AM
>To:
[hidden email]
>Subject: Re: Objective lens chromatic aberration
>- Color shift correction for colocalization
>analysis - was Glycerol Objectives - experience
>with
>
>
>
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>*****
>
>Hi all,
>
>Chromatic aberration is a measure of how the focal length (f) of a
>lens changes with wavelength. As the magnification of any optical
>system depends on the focal length of the lenses involved, of course
>this means that, to the extent that chromatic aberration is present,
>the magnification will change with wavelength. In simple optical
>systems it is visible as the pronounced red or blue fringes around
>the edges of images of high-contrast features. This is called lateral
>chromatic aberration and, if the optical system is properly aligned,
>it "should" be essentially invisible in the center of the field and
>get larger as you approach the edges. (The finding that it lines up
>in one corner but not elsewhere makes me think that there may be an
>alignment issue.)
>
>As mentioned before all refractive lenses have some chromatic
>aberration. (Reflective lenses do not). A good plan-apo, may be
>corrected for 4 wavelengths but the wavelengths chosen may vary from
>lens to lens. Between these wavelengths there will show
>wavelength-dependent shifts in focal length. The manufacturer
>endeavors to keep these shifts less than the depth of field of the
>objective so that they are not annoying to the naked eye when viewing
>multicolored specimens. In practice you will only see the effects
>when viewing planar objects such as patterns in thin metal films that
>have been evaporated onto the glass surface.
>
>Alan Boyde was the first to point out that these variations in f were
>found in even the most expensive objectives, using a tandem-disk
>scanner with Hg illumination (which has a number of intensity peaks
>in the visible) and making backscattered/reflected light images of
>planar features (surface of the slide). In fact he used this
>"problem" to make the first truly live-time 3D confocal microscope:
>Lenses were designed to have pronounced linear chromatic aberration.
>Consequently, different planes in the specimen satisfied the confocal
>conditions at different wavelengths. He then used prisms or
>diffraction elements in the oculars that were oriented so that, say,
>the red image was slightly displaced to the inside (towards the other
>eye) of the blue. Looking through the oculars, the viewer then saw a
>stereo image, obtained at full NA that was also color-coded for
>topographic height. Because confocal images could be obtained at
>several focal depths simultaneously, it was no longer necessary to
>move the specimen in z to obtain a focal series, making live-stereo
>images possible. These livetime stereo images could even be recorded
>on tape and shown to groups using special glasses
>
> From this you can see that any attempt to correct for CA must be able
>to respond to changes in magnification with wavelength as well as the
>vertical displacements between the z-location of the on-axis peak
>confocal signal vs wavelength. As noted above, there may also be
>misalignment issues.
>
>Anyone still willing to try to undertake this task should spend a lot
>of time studying the work of Cremer and Cremer in the 1980s and
>1990s. These scientists demonstrated one of the first effective
>systems of super-resolution widefield and confocal imaging as applied
>to FISH (fluorescent in-sit hybridization). They labelled one gene on
>a chromosome red, and another nearby one green. They then recorded
>the green spot and found its centroid. DItto, the red spot. Because
>these two spots were separated by wavelength, their images did not
>overlap in the recording system and therefore the distance between
>them could be measured with an accuracy that depended on the ability
>to determine the position of the centroid of each spot, rather than
>on the Abbe resolution of the optics. However, as the position of the
>centroid did depend on the CA of the optical system, they had to make
>very careful measurements of this parameter in order to apply the
>proper correction factors.
>
>Hope to see some of you in Vancouver next June!
>
>Cheers,
>
>Jim Pawley
>
>***************************************************************************
>Prof. James B. Pawley, Ph.
>608-238-3953
>21. N. Prospect Ave. Madison, WI 53726 USA
>
[hidden email]
>3D Microscopy of Living Cells Course, June 11-23, 2011, UBC, Vancouver Canada
>Info:
http://www.3dcourse.ubc.ca/ Applications due by March 15, 2011
> "If it ain't diffraction, it must be statistics." Anon.
>
>>*****
>>To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>*****
>>
>>Hi Dan,
>>The translational correction is very useful, but what if the shift varies
>>across the field?
>>Do you know if there is a plugin or a script to perform affine transform in
>>3D? The goal would be to "warp" one channel in XYZ to match the
>>second channel.
>>I have imaged 100 nm multi-color beads on our confocal (FV1000 with a
>>hand-picked 100x/1.4 Plan Super Apo lens), and besides a moderate z-shift
>>between blue and other channels (as expected), I have a lateral chromatic
>>aberration between the "DAPI" channel (405 nm ex, 430-460 em) and the longer
>>wavelength channels (green, orange, red fluorescence). So for instance, the
>>colors are colocalized in top left corner of the image, but are several
>>pixels off in XY plane in bottom right of the image. The UnwarpJ plugin for
>>ImageJ works well to correct this in individual XY images, but 3D correction
>>may be needed, since the plan correction for the blue channel (405 nm laser
>>illumination) is not as good as for the rest of the spectrum.
>>
>>
>>
>>Stan Vitha
>>Microscopy and Imaging Center
>>Texas A&M University
>>
>>
>>On Tue, 11 Jan 2011 11:23:02 +0100, Daniel James White
>><
[hidden email]> wrote:
>>
>>>*****
>>>To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>*****
>>>
>>>Hi Martin,
>>>
>>>On Jan 11, 2011, at 7:01 AM, CONFOCALMICROSCOPY automatic digest
>>>system wrote:
>>>
>>>> Date: Mon, 10 Jan 2011 17:18:29 -0600
>>>> From: Martin Wessendorf <
[hidden email]>
>>>> Subject: Re: Glycerol Objectives - experience with
>>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>> *****
>>>>
>>>> On 1/10/2011 4:32 PM, Rosemary White wrote:
>>>>
>>>>> I've found that the red and blue emission aren't quite lined up
>>>>>vertically,
>>>>> at least with our "blue" 63x objective, so
>>>>>have to cut the top one or two
>>>>> slices off a blue vertical series (depending on depth of slice), and the
> >>>> bottom slices off a red series and reassemble to get the
>>>>>emissions aligned -
>>>>> i.e. from the same depth in the tissue. The objectives do vary a bit,
>>>>> perhaps ours isn't as well corrected as some.
>>>>
>>>> Can anyone explain where the difficulty arises in correcting for axial
>>>> chromatic aberration? I've consistently seen serious axial chromatic
>>>> aberration in good quality oil objectives from one very reputable
>>>> manufacturer, between red, green and far-red (1 um off in the z-axis
>>>> between green and red, and between red and far-red; 2 um between green
>>>> and far-red). I have always heard that axial chromatic aberration was
>>>> easy to correct for and would've thought that a solution for 3-color
>>>> correction would've been found 100 years ago. However, I don't know
>>>> enough optics to understand the subtleties. Are there trade-offs to
>>>> trying to obtain good correction, besides the cost in scattering of
>>>> adding additional lens elements?
>>>
>>>Sadly, not all objective lenses were made equal.
>>>
>>>In our hands on point scanning confocals, even when you align the
>>pinhole(s) as best you can
>>>there can still be a micron of misalignment in between "green", "red" or
>>"far red",
>>>with DAPI etc. often being way off, as it comes through a different optic
> >fiber and collimator anyway (eg on an LSM 510)
>>>
>>>I have tested different lenses of the same spec from the same manufacturer,
>>>and found that they all have their own personality.
>>>Good correction is very possible, but "good" is a relative thing.
>> >
>>>I send back ones that are too "bad", and keep the "good" ones.
>>>I take z stacks of 1 micron multi colour beads to measure the
>>>remaining error
>>>(dont need sub resolution bead images for this measurement)
>>>
>>>There will always be a significant error, even in the very best
>>>"Super-dooper Mega Extra Apo-Chromat" lens you can get from any
>>manufacturer for less than 100 k dollars.
>>>They simply can not be made perfect at a reasonable cost (as explained to
>>me by a Zeiss lens guru)
>>>
>>>If you want to do precise colocalization studies at the highest optical
>>resolution (everyone does, even if they don't immediately realize it)
>>>then you simply MUST measure the error, then correct/shift the images (in
>>3D) before colocalization/correlation analysis.
>>>
>>>This can be done in 3D in ImageJ/ Fiji
>>>using nice interpolation methods from Erik Meijering, for sub pixel shifts
>>- TJ-Translate :
>>>
http://pacific.mpi-cbg.de/wiki/index.php/TransformJ>>>leading to
>>>
http://imagescience.org/meijering/software/transformj/translate.html>>>
>>>or in the great colour shift corrector of Huygens Professional (no
>>commercial interest - just a happy customer)
>>>
http://www.svi.nl/ChromaticShiftCorrector>>>
>>>or roughly in the Zeiss AIM 510 software in xy only (whole pixel shifts),
>>>
>>>The effect on the 2 channel scatterplot / 2D histogram / fluorogram is very
>>significant.
>>>A ugly poorly correlated cloud turns into tight correlated populations of
>>pixels,
>>>and the coloc. coefficients jump much higher.
>>>
>>>more info here:
>>>
http://ifn.mpi-cbg.de/wiki/ifn/index.php/Chromatic_aberration_measurement_and_correction>>>
>>>cheers
>>>
>>>Dan
>>>
>>>
>>>
>>>>
>>>> Thanks--
>>>>
>>>> Martin
>>>> --
>>>> Martin Wessendorf, Ph.D. office: (612) 626-0145
>>>> Assoc Prof, Dept Neuroscience lab: (612) 624-2991
>>>> University of Minnesota Preferred FAX: (612) 624-8118
>>>> 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009
>>>> Minneapolis, MN 55455 e-mail:
[hidden email]
>>>
>>>Dr. Daniel James White BSc. (Hons.) PhD
>>>Senior Microscopist / Image Visualisation, Processing and Analysis
>>>Light Microscopy and Image Processing Facilities
>>>Max Planck Institute of Molecular Cell Biology and Genetics
>>>Pfotenhauerstrasse 108
>>>01307 DRESDEN
>>>Germany
>>>
>>>+49 (0)15114966933 (German Mobile)
>>>+49 (0)351 210 2627 (Work phone at MPI-CBG)
>>>+49 (0)351 210 1078 (Fax MPI-CBG LMF)
>>>
>>>
http://www.bioimagexd.net BioImageXD
>>>
http://pacific.mpi-cbg.de Fiji - is just ImageJ
> >>(Batteries Included)
>>>
http://www.chalkie.org.uk Dan's Homepages
>>>
https://ifn.mpi-cbg.de Dresden Imaging
>>>Facility Network
>>>dan (at) chalkie.org.uk
>>>( white (at) mpi-cbg.de )
>
>
>--
>***************************************************************************
>Prof. James B. Pawley, Ph.
>608-238-3953
>21. N. Prospect Ave. Madison, WI 53726 USA
>
[hidden email]
>3D Microscopy of Living Cells Course, June 11-23, 2011, UBC, Vancouver Canada
>Info:
http://www.3dcourse.ubc.ca/ Applications due by March 15, 2011
> "If it ain't diffraction, it must be statistics." Anon.
>
>________________________________
>
>No virus found in this message.
>Checked by AVG - www.avg.com
>Version: 10.0.1191 / Virus Database: 1435/3342 - Release Date: 12/27/10
>Internal Virus Database is out of date.
Prof. James B. Pawley,
Ph. 608-238-3953
21. N. Prospect Ave. Madison, WI 53726 USA