Re: problem with fresh frozen eGFP sample
Posted by
Haberman, Ann on
URL: http://confocal-microscopy-list.275.s1.nabble.com/problem-with-fresh-frozen-eGFP-sample-tp593098p593099.html
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Re: problem with fresh frozen eGFP
sample
Dear Jean,
We struggled with this exact issue for quite a while before we
found the solution on the internet (embedded in someone's blog about
their frustrations!).
My understanding is that acetone or 2-methybutane achieve
fixation by dehydrating tissue. As a result, that fixation process
extracts cytoplasmic GFP. For that reason, any stains for water
soluble proteins are best performed with PFA or formalin fixed tissue.
That is the case for many proteins found within the cytoplasm.
However, GFP does not fluoresce well after formaldehyde fixation
so many investigators have included an additional anti-GFP stain to
detect the presence of GFP that no longer can fluoresce! Rockland
makes an anti-GFP that also detects YFP.
best,
Ann
Content-type: text/plain;
charset=utf-8
Content-language:
Hi all,
I have an animal that is Tg eGFP for a neurotransmitter in
astrocytes. It shows up very well in brain and spinal cord after
perfusing with 4%PFA. I'm trying to Laser capture some cells for
RNA work but having trouble visualizing the cells in the fresh frozen
tissue. I'm doing cervical dislocation, the disection is under
1min., then placing the tissue into 2-Methylbuane that has been cooled
and surrounded by dry ice. The tissue is then cut in a cryostat,
8um section placed onto a slide. I then visualize using either
laser or mercury bulb (with 488 filter) and the cells look abstract or
green everywhere instead of dark background with green astrocytes
(star shaped cells). I have cut frozen sections from PFA
perfused animals several times and there is no problem with the way
the cells look. Do you think I need to hydrate the slide to
visualize the cells better? Is the freezing method faulty?
Any suggestions?
Thanks for your help.
Jean
Jean Brennan
Research Specialist
Rothstein lab, Dept of Neurology
Johns Hopkins University
600 N. Wolfe St., Meyer 6-174
Baltimore, MD 21287
410-614-4119; FAX: 410-955-0672
[hidden email]
Content-type: text/html; charset=utf-8
Content-language:
Hi all,
I have an animal that is Tg eGFP for
a neurotransmitter in astrocytes. It shows up very well in brain
and spinal cord after perfusing with 4%PFA. I'm trying to
Laser capture some cells for RNA work but having trouble visualizing
the cells in the fresh frozen tissue. I'm doing
cervical dislocation, the disection is under 1min., then placing the
tissue into 2-Methylbuane that has been cooled and surrounded by dry
ice. The tissue is then cut in a cryostat, 8um section placed
onto a slide. I then visualize using either laser or mercury
bulb (with 488 filter) and the cells look abstract or green everywhere
instead of dark background with green astrocytes (star shaped cells).
I have cut frozen sections from PFA perfused animals several times and
there is no problem with the way the cells look. Do you think I
need to hydrate the slide to visualize the
cells better? Is the freezing method faulty? Any
suggestions?
Thanks for your help.
Jean
Jean Brennan
Research Specialist
Rothstein lab, Dept of Neurology
Johns Hopkins University
600 N. Wolfe St., Meyer 6-174
Baltimore, MD 21287
410-614-4119; FAX: 410-955-0672
[hidden email]
--
Ann Haberman, PhD
Department of Laboratory Medicine
Yale University School of Medicine
1 Gilbert St.
TAC S541
New Haven, CT 06510
203-785-7349
203-785-5415 (fax)
[hidden email]