Re: problem with fresh frozen eGFP sample
Posted by
Ian Dobbie-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/problem-with-fresh-frozen-eGFP-sample-tp593098p593110.html
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocalJean Brennan <
[hidden email]> writes:
> Hi all,
>
> I have an animal that is Tg eGFP for a neurotransmitter in astrocytes. It
> shows up very well in brain and spinal cord after perfusing with 4%PFA. I'm
> trying to Laser capture some cells for RNA work but having trouble visualizing
> the cells in the fresh frozen tissue. I'm doing cervical dislocation, the
> disection is under 1min., then placing the tissue into 2-Methylbuane that has
> been cooled and surrounded by dry ice. The tissue is then cut in a cryostat,
GFP fluorescence is completely dependant upon structural water within
the beta barrel[1]. I'm not sure how strongly you are dehydrating your
sample, but this could cause you to loose you GFP signal.
Ian
[1] see Tsien, RY. THE GREEN FLUORESCENT PROTEIN
Annu. Rev. Biochem. 1998.67:509-544.