Posted by
Stephen Cody on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Strange-artifact-in-confocal-Z-stack-tp593216p593228.html
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocalD'day Christophe,
I think that it is non-specific fluorescence on the outside of your tissue section. Did you use an antibody label? If so it may represent non-specific binding to the outside of the section. If it was a lipophilic retrograde label such as DiI, these can wash out of the membrane when in glycerol based mountants, and it may have become trapped between the section and the coverglass.
The curved nature of the artefact is probably as Paul suggests, because of the use of a non plan objective. Further evidence that you may be looking at the outside of the tissue section is that there appears to be knife chatter marks visible in panel 5.
Cheers
Stephen H. Cody
Microscopy Manager
Central Resource for Advanced Microscopy
Ludwig Institute for Cancer Research
PO Box 2008 Royal Melbourne Hospital
Victoria, 3050
Australia
Tel: 61 3 9341 3155 Fax: 61 3 9341 3104
email:
[hidden email]
www.ludwig.edu.au/labs/confocal.html
www.ludwig.edu.au/confocal
Tip: Learn how to receive reminders about you microscope booking:
www.ludwig.edu.au/confocal/Local/Booking_Hint.htm
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Christophe Leterrier
Sent: Thursday, 29 May 2008 6:30 PM
To:
[hidden email]
Subject: Strange artifact in confocal Z-stack
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocalDear all,
One of my colleagues brought me a Z-stack today that has a circular
pattern superimposed on his sample. It is taken with a Leica SPE
confocal, pinhole at 1 Airy unit, I guess the Z step is about 0.2 µm.
If you go down the stack trough the cell toward the glass (cells on
glass coverslip observed on an upright microscope), you see a circular
ring halo appearing that lowers in radius up to a circular spot at the
center of the field. A picture is worth a thousand words, so please
check the problem here :
http://ftp.espci.fr/incoming/christo/Montage.jpgI've seen reflections from the coverslip before as a high background
fluorescence appearing when you reach the glass surface, but I wonder
if this is reflection or a defect in the objective, or something else.
Sis anyone have a similar problem ?
Thanks a lot for your help,
Christophe
Christophe Leterrier
Postdoc
INSERM UMR641 Neurobiology of ionic channels
IFR Jean Roche - Mediterranee University
Marseille, France
This communication is intended only for the named recipient and may contain information that is confidential, legally privileged or subject to copyright; the Ludwig Institute for Cancer Research does not waiver any rights if you have received this communication in error.
The views expressed in this communication are those of the sender and do not necessarily reflect the views of the Ludwig Institute for Cancer Research.