Hi everybody,
 
I don't understand why the dynamic range of a CCD is defined as the full-well capacity divided by the camera noise, and not minus the camera noise.
 
Thanks a lot for your help,
 
Laurent.
 
___________________________________________
Laurent Gelman, PhD
Friedrich Miescher Institut
Facility for Advanced Imaging and Microscopy
WRO 1066.2.16
Maulbeerstrasse 66
CH-4058 Basel
Tel.: 061 696 43 38 / Cell phone: 079 618 73 69
www.fmi.ch
 
 

Dear Laurent.

Actually it is a misleading way to define dynamic range.  The idea is that the read noise gives you the error in the intensity of any measurement. So this definition of dynamic range implies that it is a measure of the number of separable intensity levels that might be visible in the data.

The problem is that, once the signal level exceeds the square of the read noise (in electron/pixel) the ability of discriminate different signal levels becomes limited  more by Poisson Noise than by the "dynamic range".

So, in sum, it is just a convention that works pretty well to separate better CCDs from worse.

For microscopists however, the read noise is a more useful measure because we seldom have specimens bright enough to produce a "full-well" signal.

Jim P.
-- 