Re: Ask for help about LSM 510

Posted by Melinda Kay Frame on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Ask-for-help-about-LSM-510-tp593440p593442.html

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Yi-

I have experienced the "Failed to connect to real time controller" problem
twice in the past two months on my Zeiss 510 Meta.

There is a secondary computer or controller in the large module that
contains the lasers and electronics.  This secondary computer is connected
to the Windows computer by an internet cable.  When you start the Windows
program, you should see at the bottom right of the monitor on one of the
icons a message indicating "Real Time Connection/Speed 1Gbps/Status
Connected"

The first time I experienced the "Failed to connect to real time
controller", the icon on the Windows screen also indicated that the Real
Time Connection was Not Connected.  That time I could not get the system to
work without a service call.

The second time, the icon on the Windows screen indicated that the Real Time
Connection was still Connected at 1Gbps.  If I shut down the system for
30-60 minutes (5 or 10 minutes wouldn't work), then I could get the system
back up and working.  It was as though there was a heating issue or
something.

In both cases, the final solution was for Zeiss to replace that secondary
computer in the electronics module.  The system worked fine after that.

Hope this helps-
Melinda

Melinda Frame, PhD
Michigan State University
Center for Advanced Microscopy
B7 CIPS
East Lansing, MI  48824
517-432-2327
[hidden email]

Yi Zheng writes:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
>
> Hi everyone!
>
> Here is an emergency.
>
> I am doing my data analysis on zeiss LSM 510 while the whole system is
> runing. I mean all laser are on and I was collecting data.
>
> The software was down suddenly.
> So I quit it and tried to restart it.
> Then I got the message like "Fail to connect to realtime controller. If the
> system is running open the door "reset RTC" on the front of the electronics
> rack. press the "Reset" button and wait 30 seconds.
> But the problem is I can not find any door or reset button behind it.
>
> I called Zeiss, but their system is down too!
>
> Is there anyone here can help?
>