Re: Ask for help about LSM 510
Posted by
Zoltan on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Ask-for-help-about-LSM-510-tp593440p593450.html
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
As far as I can remember two years back when I was still working on a Zeiss system (LSM 5), there is indeed a door just above the floor level on the vertical side of the main controller unit. It should be facing you if you are sitting in front of the computer screen. The reset button is behind it. I never tried to do analysis data while collecting data, and I'm still not...
Hope it'll work.
Zoltan
On Thu, Jul 17, 2008 at 10:35 PM, Yi Zheng <
[hidden email]> wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hi everyone!
Here is an emergency.
I am doing my data analysis on zeiss LSM 510 while the whole system is
runing. I mean all laser are on and I was collecting data.
The software was down suddenly.
So I quit it and tried to restart it.
Then I got the message like "Fail to connect to realtime controller. If the
system is running open the door "reset RTC" on the front of the electronics
rack. press the "Reset" button and wait 30 seconds.
But the problem is I can not find any door or reset button behind it.
I called Zeiss, but their system is down too!
Is there anyone here can help?
--
--
Zoltan Cseresnyes
Facility manager, Imaging Suite