I would say that the size of your optical section is entirely
dependant on the NA of your objective. What happens when you open your
pinhole is that you let photons coming from "outside" contribute to
image formation. Of course, you also get back difracted light coming
from your focal plane that you otherwise you would have rejected.
The end effect is more a blurring than a "thickening".
the simplest solution is to get a infra red pulsed laser! No need of
pinhole with multiphoton!