Dear Craig,
We use it for ‘simple’ confocal microscopy. We are
located in a city, close to a railway station, we have some centrifuges
running next door, our facility is on the 2nd floor (not my idea to
build it there), so we never reached anywhere near the resolution specs until
we switched to active dumping.
jens
From: Confocal Microscopy
List [mailto:[hidden email]] On Behalf Of Craig Brideau
Sent: Monday, July 21, 2008 11:28 PM
To: [hidden email]
Subject: Re: help for optical table choice
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
If I might ask, what
application were you using that you required active damping? I haven't
heard of this a whole lot outside of AFM and the semiconductor industry,
although I see it is becoming more popular for highly sensitive applications.
Craig
On Sat, Jul 19, 2008 at 5:05 AM, Rietdorf, Jens
<[hidden email]> wrote:
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello Jean-Pierre,
we got excellent results with the TS series tables from
http://www.hwlscientific.com/
(no commercial interest) in an environment
where all kinds of passive dumping have failed. You may want to include
these into your evaluation.
Best, jens
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On
Behalf Of Jean-Pierre CLAMME
Sent: Friday, July 18, 2008 12:20 AM
To: [hidden email]
Subject: help for optical table choice
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hi,
I'm shopping for an optical table for a new confocal / 2photon
microscope
and FCS setup and I was wondering what kind of grade I need.
Indeed for the 2 photon microscope that we have here ((the new one is
for
another place) , we have a TMC 784 series (maximum dry damping) table.
I
saw somewhere that this table is equivalent to the RS 4000 series from
NEWPORT.
So I was wondering if this grade is really necessary for a confocal
microscope and FCS system. Or if as long as I have pneumatic legs, a
RS2000
or 783 (standard damping) would be fine ?
Thank you in advance,
JP