Great advice. Thank you.
Kathy
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Yevgeniy Romin
Sent: Thursday, February 10, 2011 5:54 PM
To:
[hidden email]
Subject: Re: tokai hit chamber
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Hi Kathy
We have several Tokai Hit icubators in our facility, they all dry out within 24 hours, the company seems to be aware of it and they suggest to put in more water every 12 hours, so that is what we do. Wet Kimwipes also help. Or wet cotton balls.
________________________________________
From: Confocal Microscopy List [
[hidden email]] On Behalf Of Kathryn Spencer [
[hidden email]]
Sent: Thursday, February 10, 2011 7:17 PM
To:
[hidden email]
Subject: Re: tokai hit chamber
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Thanks Ippei;
We are imaging 2-3 days, and the reservoir is drying out. Short term, it does seem OK. I'm just unsure about why it is drying out, if we are just imaging much longer than normal. If we put in wet Kimwipes, everything seems fine. Somehow, we need to add water during our imaging.
Kathy
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Ippei Kotera
Sent: Thursday, February 10, 2011 3:56 PM
To:
[hidden email]
Subject: Re: tokai hit chamber
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Hi Kathy,
Tokai Hit System I had been using was equipped with a doughnut-shaped
water chamber and heated lid along with CO2 and heat supplier. I have
done live imaging up to 5 hours with no humidity problem. What kind of
humidity issues do you have? The water chamber was simple enough that
nothing could go wrong, so it seemed.
Ippei
--
Ippei Kotera, Ph.D.
Tanz CRND
Tanz Neuroscience Building, Room 221
University of Toronto
6 Queen's Park Crescent West
Toronto, Ontario M5S 3H2
Phone: (416)978-2503
[hidden email]
=====================================================================
Please note that this e-mail and any files transmitted with it may be
privileged, confidential, and protected from disclosure under
applicable law. If the reader of this message is not the intended
recipient, or an employee or agent responsible for delivering this
message to the intended recipient, you are hereby notified that any
reading, dissemination, distribution, copying, or other use of this
communication or any of its attachments is strictly prohibited. If
you have received this communication in error, please notify the
sender immediately by replying to this message and deleting this
message, any attachments, and all copies and backups from your
computer.