Hi Kathy,

Hi Kathy,

If you have a 5% CO2 gas flow into the chamber, have you tried bubbling it through a small reservoir of distilled water?  Tank gas is quite dry.  A while back I used that trick in a jerryrigged unit for 24-h. imaging.  We have a Tokai-Hit now but have not yet needed it for very long-term imaging.  

cheers,


Tim Feinstein


On Feb 10, 2011, at 7:17 PM, Kathryn Spencer wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Thanks Ippei;
> We are imaging 2-3 days, and the reservoir is drying out. Short term, it does seem OK. I'm just unsure about why it is drying out, if we are just imaging much longer than normal. If we put in wet Kimwipes, everything seems fine. Somehow, we need to add water during our imaging.
> Kathy
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ippei Kotera
> Sent: Thursday, February 10, 2011 3:56 PM
> To: [hidden email]
> Subject: Re: tokai hit chamber
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Kathy,
>
> Tokai Hit System I had been using was equipped with a doughnut-shaped
> water chamber and heated lid along with CO2 and heat supplier. I have
> done live imaging up to 5 hours with no humidity problem. What kind of
> humidity issues do you have? The water chamber was simple enough that
> nothing could go wrong, so it seemed.
>
> Ippei
>
> --