To elaborate a bit on that, why isn't it possible to go beyond the
diffraction limit with deconvolution ? I don't see the theoretical reason,
in a ideal case (subresolutive beads sufficiently spaced). Of course in a
real sample, S/N issues and fluorophore density impose a limit on the
precision, but why would this limit precisely be at the diffraction limit ?

Christophe

On Wed, Apr 13, 2011 at 09:44, Andreas Bruckbauer <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
>
> Hi Dan, Brian
>
> thanks for the comments, i know that for real live samples everything what
> exceeds the resolution limit of the system is an artefact, however for
> sparsely separated beads i would expect that i can determine their position
> with high accuracy only limited by the signal to noise ratio. This is
> exactly what is done in localisation microscopy (PALM, STORM...). One
> approach is to calculate the cross correlation between the measured 3D PSF
> and the 3D image (see e.g. A. G. York et al. Nature Methods 8, 2011, 327).
> While this is not exactly what deconvolution does, it is at least related to
> deconvolution.
>
> So i was wondering if this special case, a limited number of well resolved
> beads which are small enough to qualify as point objects, could tell us how
> good the deconvolution method actually works, or in the case of SIM, what
> the image restauration actually does. At least theoretically, the size of
> the beads (110 nm in my case) should not be a problem as by using the same
> beads to measure the PSF we trick the deconvolution into thinking that they
> are point objects. This is similar to the obove mentioned localisation
> microscopy which does not depend much on optical resolution or pixel size.
>
> best wishes
>
> Andreas
>
>
>
>
>
>
>
>
>
>
>
>
> -----Original Message-----
> From: Daniel James White <[hidden email]>
> To: [hidden email]
> Sent: Mon, 11 Apr 2011 10:53
> Subject: Re: Deconvolution of 3D SIM data
>
>
> *****
>
>
>
>
>
> To join, leave or search the confocal microscopy listserv, go to:
>
>
>
>
>
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
>
>
>
>
> *****
>
>
>
>
>
>
>
>
>
>
>
> Hi Andreas,
>
>
>
>
>
>
>
>
>
>
>
> On Apr 9, 2011, at 7:02 AM, CONFOCALMICROSCOPY automatic digest system
> wrote:
>
>
>
>
>
>
>
>
>
>
>
> > Date:    Fri, 8 Apr 2011 11:48:46 -0400
>
>
>
>
>
> > From:    Andreas Bruckbauer <[hidden email]>
>
>
>
>
>
> > Subject: Re: Deconvolution of 3D SIM data
>
>
>
>
>
> >
>
>
>
>
>
> >
>
>
>
>
>
> > When i generate a PSF for deconvolution using suitable beads, then image=
>
>
>
>
>
> > the same beads again and deconvolve the image, i would expect to get
> real=
>
>
>
>
>
> > ly tiny dots. See e.g. http://www.svi.nl/BeadsDeconvolutionExample=20
>
>
>
>
>
>
>
>
>
>
>
> real tiny? The images of sub resolution objects dont get very much smaller
> after
>
>
>
>
>
> deconvolution...
>
>
>
>
>
> what you really get is much higher contrast, so the features look "sharper"
>
>
>
>
>
> You also get a little bit more resolution.. but thats not really the main
> point.
>
>
>
>
>
>
>
>
>
>
>
> Its really about contrast.
>
>
>
>
>
>
>
>
>
>
>
> > I would expect something similar for the OMX when the reconstruction is
> pe=
>
>
>
>
>
> > rfect and includes proper deconvolution. However we get larger features,=
>
>
>
>
>
> > still within the expected OMX resolution 120 nm in width and 300 nm in
> z,=
>
>
>
>
>
> > but no dots.
>
>
>
>
>
>
>
>
>
>
>
> No reconstruction or deconvolution will give images that contain features
> that
>
>
>
>
>
> are smaller
>
>
>
>
>
> than the resolution limit of the system.
>
>
>
>
>
>
>
>
>
>
>
> What do you mean by "dots" ?
>
>
>
>
>
> Single 40 nm pixels?
>
>
>
>
>
>
>
>
>
>
>
> In the OMX it's twice the conventional resolution,
>
>
>
>
>
> so you cant get objects that appear smaller than 120 nm.
>
>
>
>
>
> (actually... API say the OMX system sometimes seems to outperform theory
>
>
>
>
>
> slightly... but only a bit)
>
>
>
>
>
>
>
>
>
>
>
> You should not an must not get single pixel object with the reconstructed
> pixel
>
>
>
>
>
> spacing of 40 nm (thats the spacing we have with the EM CCDs after 2x
> resolution
>
>
>
>
>
> increase from the physical 80 nm pixel spacing)
>
>
>
>
>
>
>
>
>
>
>
> If you get smaller objects, its wrong, and an artifact. No?
>
>
>
>
>
>
>
>
>
>
>
> > I think there is still some improvement possible either in=
>
>
>
>
>
> > the setup of the instrument or the software.
>
>
>
>
>
>
>
>
>
>
>
> Actually, in my hands it seems to do what theory predicts and no more.
>
>
>
>
>
> There is no good reason to expect it to make smaller objects than 120 nm.
>
>
>
>
>
>
>
>
>
>
>
> Certainly the careful alignment of the optics and the highest quality
> objective
>
>
>
>
>
> lens are critical,
>
>
>
>
>
> as is the measurement of suitable SIM PSFs  and careful
> calibration/measurement
>
>
>
>
>
> of the parameters for the reconstruction...
>
>
>
>
>
> the angles and the phases.
>
>
>
>
>
>
>
>
>
>
>
> So, other than making it faster and more tolerant to sub optimal input
> data,
>
>
>
>
>
> I dont think there is any improvement to be made in resolution.. so I dont
> see
>
>
>
>
>
> how you think there is room for resolution improvement?
>
>
>
>
>
> Maybe you can share your ideas?
>
>
>
>
>
>
>
>
>
>
>
> cheers
>
>
>
>
>
>
>
>
>
>
>
> Dan
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
>
> >
>
>
>
>
>
> > best wishes
>
>
>
>
>
> >
>
>
>
>
>
> > Andreas
>
>
>
>
>
> >
>
>
>
>
>
> >
>
>
>
>
>
> > =20
>
>
>
>
>
> >
>
>
>
>
>
> > =20
>
>
>
>
>
>
>
>
>
>
>
> Dr. Daniel James White BSc. (Hons.) PhD
>
>
>
>
>
> Senior Microscopist / Image Visualisation, Processing and Analysis
>
>
>
>
>
> Light Microscopy and Image Processing Facilities
>
>
>
>
>
> Max Planck Institute of Molecular Cell Biology and Genetics
>
>
>
>
>
> Pfotenhauerstrasse 108
>
>
>
>
>
> 01307 DRESDEN
>
>
>
>
>
> Germany
>
>
>
>
>
>
>
>
>
>
>
> +49 (0)15114966933 (German Mobile)
>
>
>
>
>
> +49 (0)351 210 2627 (Work phone at MPI-CBG)
>
>
>
>
>
> +49 (0)351 210 1078 (Fax MPI-CBG LMF)
>
>
>
>
>
>
>
>
>
>
>
> http://www.bioimagexd.net   BioImageXD
>
>
>
>
>
> http://pacific.mpi-cbg.de       Fiji -  is just ImageJ (Batteries
> Included)
>
>
>
>
>
> http://www.chalkie.org.uk       Dan's Homepages
>
>
>
>
>
> https://ifn.mpi-cbg.de          Dresden Imaging Facility Network
>
>
>
>
>
> dan (at) chalkie.org.uk
>
>
>
>
>
> ( white (at) mpi-cbg.de )
>
>
>
>
>
>
>
>
>