Abbe considered rays diffracted by two points on the sample.  The points
will be resolved if their diffracted rays can enter the objective.  This
can only apply to  a specimen illuminated from an external source.  In
fluorescence each point emits light and this is totally incoherent with
respect to light from another point.   There is no diffraction at the
sample so Abbe's calculation cannot be applied.  Rayleigh's criterion,
based on how the optics turn the image of a point into a disk (the Airy
disk) does apply.

 

 
Guy

 

Sponsor my next half-marathon on May 15th

There's a special reason - find it out at

http://www.everydayhero.com.au/Guy_Cox_4846
<http://www.everydayhero.com.au/Guy_Cox_4846>

______________________________________________

Associate Professor Guy Cox, MA, DPhil(Oxon)

Australian Centre for Microscopy & Microanalysis,

Madsen Building F09, University of Sydney, NSW 2006

 

Phone +61 2 9351 3176     Fax +61 2 9351 7682

             Mobile 0413 281 861

______________________________________________

      http://www.guycox.net <http://www.guycox.net/>

 

From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Martin Wessendorf
Sent: Thursday, 14 April 2011 12:10 AM
To: [hidden email]
Subject: Re: Deconvolution of 3D SIM data

 

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

On 4/13/2011 7:33 AM, Guy Cox wrote:

> Abbe's limit,
> on the other hand, is non-negotiable, but it doesn't apply in
> fluorescence.

Guy--can you elaborate a bit on your statement that, "it doesn't apply
in fluorescence"?  Are you talking about STED, or talking in more
general terms?

Thanks--

Martin
--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455                    e-mail: [hidden email]

________________________________

No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1209 / Virus Database: 1500/3569 - Release Date: 04/12/11