I'm afraid the idea is MUCH older than 2008. I recall reading a paper
limit was not a real limit at all. I believe they were using bacteria
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> I think this is the conclusion of a paper by Sripad Ram (PNAS 103,
> 2006, 4457), but i think in this case you have to know the number of
> objects you are looking at.
>
> best wishes
>
> Andreas
>
>
>
>
>
>
>
>
>
>
> -----Original Message-----
> From: Martin Wessendorf <
[hidden email]>
> To:
[hidden email]
> Sent: Thu, 14 Apr 2011 4:41
> Subject: Re: Deconvolution of 3D SIM data
>
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> On 4/13/2011 5:28 PM, Guy Cox wrote:
>
>> Abbe considered rays diffracted by two points on the sample. The
>> points
>> will be resolved if their diffracted rays can enter the objective.
>> This
>> can only apply to a specimen illuminated from an external source.
>> In
>> fluorescence each point emits light and this is totally incoherent
>> with
>> respect to light from another point. There is no diffraction at the
>> sample so Abbe's calculation cannot be applied. Rayleigh's
>> criterion,
>> based on how the optics turn the image of a point into a disk (the
>> Airy
>> disk) does apply.
>
> This is (for me!) a very intuitive explanation, but it suggests that
> with fluorescence, arbitrarily small resolution can be attained
> given sufficiently high s/n. That sounds something like what you
> said in your earlier posting, except for the phrase "arbitrarily
> small".
>
> Is that correct? If not, what is the absolute limit of resolution
> in fluorescence?
>
> Martin
>