> the 1 photon PSF. This has been described in numerous texts.
Mark, I am glad to hear that. I didn't come across those texts, though.
theoretically achievable, starting with Ri mismatch. But again, if you
>
> Cheers
>
>
>
> On 15/04/2011, at 5:52 AM, Steffen Dietzel wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> Hi everybody,
>>
>> somewhat related to the ongoing discussion on resolution, I came
>> across a puzzle today concerning the resolution of multi-photon
>> microscopy.
>>
>> I measured the resolution of our third harmonic generation (THG)
>> microscope and surprisingly I came up with a full width half maximum
>> (FWHM) slightly better than theory allows. Seems the most likely
>> explanation is I applied the wrong theory. But which one is the
>> correct one?
>>
>> The experiment:
>> THG with 1275 nm, Objective 0.95 NA (water, 20x), beads 60 nm in 2%
>> agarose, voxel size 0.136 x 0.136 x 0.5 µm. Result: FWHM ~0.7 µm (for
>> both, forward and backward THG)
>>
>> Assuming that for multi-photon point-scanners only the excitation
>> wavelength is relevant, I used 1275 nm for the theory (Rayleigh):
>> r=0.61λ/NA = 0.82 µm
>>
>> So, the measured resolution is one pixel better than the theoretical
>> limit. You don't get that lucky every day ;-)
>>
>> Possibilities I have considered:
>> - I messed up the experiment. I wouldn't know, however, how I could
>> get a better result by messing up.
>> - Microscope settings are wrong (wrong pixel size). Possible of course
>> but not very likely.
>> - Rayleigh does not apply to multi-photon, I overlooked something. If
>> so, please help out.
>> - THG requires 3 photons to take place. Maybe the photon density is
>> low enough in the outer areas of the PSF so that signal generation is
>> limited to inner areas of the PSF? (Now that would be really
>> interesting from an academic point of view, since it would mean you
>> could squeeze the size of the excitation spot relative to the
>> wavelength with 4, 5, etc. photon effects. Although it probably
>> wouldn't do much good for practical purposes since you would have to
>> start with long wavelengths to end up with a visible (=easy
>> detectable) signal.)
>>
>>
>> Any ideas? Could people share measured FWHMs from their multi-photon
>> setup? Maybe even from a 3 photon process?
>>
>> Steffen
>>
>>
>> --
>> ------------------------------------------------------------