Re: Deconvolution of 3D SIM data
Posted by
Ian Dobbie on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Deconvolution-of-3D-SIM-data-tp6251420p6281660.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Nuno Moreno <
[hidden email]> writes:
> Therefore, I would be very, very skeptical either on any deconvolution
> and quantification using SIM.
Actually if your data has high S/N then quantitation should be
possible. For 2D super-resolution with SIM you have an over specified
linear problem, 9 images to give 4x the data (double resolution in
x&y). With 3D-SIM, as produced by OMX scopes, the problem is similar, 15
images producing 8x as much information (double res in x,y & z). Go look
at the original Gustafson papers for the details.
The issue is getting high enough S/N, which in practical terms means low
enough bleaching while taking those 100+ images.
Ian
--
Ian Dobbie
Micron Imaging Facility Manager,
Biochemistry,
University of Oxford,
South Parks Road,
Oxford
OX1 3QU
Tel: 01865 613323
Email:
[hidden email]