Re: FocalCheck test slide: Vendor Update

Posted by John Oreopoulos on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Re-FocalCheck-test-slide-Vendor-Update-tp6270650p6284831.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I asked this same question about the mounting medium a few months ago. Here is the response I got from Molecular Probes:

***********************
John:
 
The mounting medium is Saint Gobain Optical Cement.  According to my notes, this has a refractive index of 1.52 when cured.
 
The spectra for the TetraSpeck green dye are the same as our standard yellow/green FluoSpheres: http://www.invitrogen.com/site/us/en/home/support/Product-Technical-Resources/Product-Spectra.8811h2o.html
 
The spectra for the Tetraspeck dark red  dye are the same as our standard dark red FluoSpheres: http://www.invitrogen.com/site/us/en/home/support/Product-Technical-Resources/Product-Spectra.8807h2o.html
 
The blue and orange dyes are different from the standard FluoSpheres dyes so I don't have data for those at present.  I'll work on getting it but may take a few weeks.
 
Best Regards,
 
iain
***********************


John Oreopoulos
Research Assistant
Spectral Applied Research
Richmond Hill, Ontario
Canada
www.spectral.ca


On 2011-04-18, at 1:49 PM, Dan White wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Mike and Pete
>
> On Apr 15, 2011, at 7:02 AM, CONFOCALMICROSCOPY automatic digest system wrote:
>
>>
>> Date:    Thu, 14 Apr 2011 10:27:38 +0200
>> From:    Peter Gabriel Pitrone <[hidden email]>
>> Subject: Re: FocalCheck test slide: Vendor Update
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Dear Mike,
>>
>> This is great news! Thanks for listening to your costumers wishes.
>
> Halleluja!
>
>>
>> I still am of the opinion that each bead size should be on it's own
>> slide. The "one slide fits all" method (with 6 positions in a straight
>> line) needs to be discarded, because it is impossible to even reach the
>> last two positions.
>
> yes, and the beads need to be in the middle of the slide to be visible on an OMX
> or other system / stage with restricted stage travel.
>
> Also.. what exactly is the refractive index of the mounting medium,
> and what exactly is it anyway?
>
> Can we get 50 nm diameter bead slides for super resolution structured illumination microscope systems like the omx?
>
> 100 or 170 nm are not small enough....
>
> cheers
>
> Dan
>
>>
>> Thanks again!
>>
>> and best regards,
>>
>> Pete
>>
>>
>> On Wed, 2011-04-13 at 16:01 -0500, Mike Ignatius wrote:
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Dear Listserve members,
>>>
>>> We are pleased to announce that we were able to improve the mounted
>>> tetraspec beads products described in this colorful email string from over a
>>> year ago (with some comments extending back to 2005).    
>>>
>>> Tertraspec beads are now mounted on the #1.5 coverslips, not the slides.  
>>>
>>> It took some doing - overspray from the mask used to find the bead focal
>>> plane was our biggest unanticipated issue that delayed this so much.  Many
>>> apologies to anyone out there that remained on backorder.  But the product,
>>> T14792 and other bead mounted slides, like our intensity and size bead series,
>>> F36909, are now on the coverslip.  
>>>
>>> Frankly they look terrific - wish I could post an image.
>>>
>>> Kind Regards,
>>>
>>> Mike Ignatius
>>> Molecular Probes/LifeTechnologies
>>>
>>> On Fri, 26 Mar 2010 17:31:10 -0400, Ignatius, Mike
>>> <[hidden email]> wrote:
>>>
>>>> Dear Listservors,
>>>>
>>>> Thank you all very much for your contributions to this discussion.
>>>>
>>>> We have always used #1.5 coverslips(which range from 160 to 190
>>> microns).  All biological objectives (except those with correction collars) are
>>> designed for these. Going to #1 coverslips would only narrow things by 30
>>> microns at best and defeat the optical correction factor.
>>>>
>>>> We continue to encourage do-it-yourselfers (DYI) slides as others have
>>> mentioned, and include protocols to do this.  We are very generous in the
>>> quantities of beads shipped - you can make dozens to hundreds of slides from
>>> one vial. However we developed the multiple bead slides as a cost savings to
>>> the customer needing more than one bead type.  Buying 10 bead preps is 2-5x
>>> what one of the prepared slides will cost.  
>>>>
>>>> Unfortunately, the prepared slides are a quite a challenges to make in
>>> quantity.  We have strict QC standards - our scrap rate and thus costs are
>>> high.  We haven't implemented a QC step with oil objectives because of the
>>> damage to the slide that would entail.  As mentioned in an earlier post we are
>>> addressing this going forward (first up mounting on coverslips).  
>>>>
>>>> We will refund anyone that is unable to see their beads on our slides of
>>> course.  Just call our Tech Support for a credit or replacement, at 800-955-
>>> 6288, then selection 2, then selection 4 to request this.
>>>>
>>>> Regards,
>>>>
>>>> Mike Ignatius
>>>>
>>>> Molecular Probes/Lifetechnologies
>>>>
>>>>
>>>>
>>>> -----Original Message-----
>>>> From: Confocal Microscopy List
>>> [mailto:[hidden email]] On Behalf Of Peter Pitrone
>>>> Sent: Friday, March 26, 2010 6:03 AM
>>>> To: [hidden email]
>>>> Subject: Re: FocalCheck test slide
>>>>
>>>> They sell these "test slides" for hundreds of dollars, yet they don't take the
>>> time to make them right?!?! They should be made with correct coverslips
>>> chosen for their thickness, deposited on them and then mounted on the glass
>>> slide... I see that there is a disconnect some where, how much does it cost
>>> them to produce these slides? 10 bucks maybe...
>>>>
>>>> Pete
>>>>
>>>> On Mar 26, 2010, at 12:25 PM, Guy Cox wrote:
>>>>
>>>>> No, no, it IS the problem!  As Mike Ignatius explained, MP put the beads on
>>> the slide, not the coverslip, then add mountant and then the coverslip.  If you
>>> are using a #1.5 coverslip you need to put the beads directly on the
>>> coverslip.  With the beads on the slide the extra thickness of the mountant
>>> needs to be corrected for by using a thinner coverslip - #1 or #0 - which
>>> must be found by trial and error.  (But since I guess they are using very
>>> reproducible conditions they only need to do the test once).  At least it seems
>>> they have realized they have a problem.  
>>>>>
>>>>>                                     Guy
>>>>>
>>>>> -----Original Message-----
>>>>> From: Confocal Microscopy List
>>> [mailto:[hidden email]] On Behalf Of MICROSCOPIA
>>> CONFOCAL y CCD
>>>>> Sent: Thursday, 25 March 2010 10:15 PM
>>>>> To: [hidden email]
>>>>> Subject: Re: FocalCheck test slide
>>>>>
>>>>>> Thanks a lot for your answers.
>>>>>
>>>>>> Of course I put the slide upside down when using
>>>>>> inverted microscope and I asked for Molecular
>>>>>> probes about coverslip and they answered me:
>>>>>> "The coverslip thickness for the FocalCheck
>>>>>> prepared slides is 0.13-0.16 mm" It´s no ideal
>>>>>> but I think that this is not the problem.
>>>>>
>>>>> M. Teresa
>>>>>
>>>>> SERVICIO de MICROSCOPIA CONFOCAL y CCD
>>>>> Mª Teresa Seisdedos Domínguez
>>>>> Oscar Hidalgo Blanco
>>>>> Amadeo Cazaña Soto
>>>>> [hidden email]
>>>>> Centro Investigaciones Biológicas (CIB) CSIC
>>>>> C/Ramiro de Maeztu, 9
>>>>> 28040 Madrid
>>>>> Phone: + 34-91 8373112 ext.4401
>>>>> Fax: + 34-91 536 04 32  
>>
>> --
>> Peter Gabriel Pitrone
>> Microscope Specialist
>> Max Planck Institute
>> for Molecular Cell
>> Biology and Genetics
>> Pfotenhauerstr. 108
>> 01307 Dresden
>> Germany
>
> Dr. Daniel James White BSc. (Hons.) PhD
> Senior Microscopist / Image Processing and Analysis
> Light Microscopy Facility
> Max Planck Institute of Molecular Cell Biology and Genetics
> Pfotenhauerstrasse 108
> 01307 DRESDEN
> Germany
>
> +49 (0)15114966933 (German Mobile)
> +49  (0)351 210 2627 (Work phone at MPI-CBG)
> +49  (0)351 210 1078 (Fax MPI-CBG LMF)
>
> http://www.bioimagexd.net                BioImageXD
> http://pacific.mpi-cbg.de                Fiji (is just ImageJ - batteries included)
> http://www.chalkie.org.uk
> [hidden email]
> ( [hidden email] )