Hi David,

For this kind of application MP has no advantage other than no
out-of-focus photobleaching. CLSM is a better solution.

With regards,
Vladimir Ghukasyan
Confocal and Multiphoton Imaging Facility
Neuroscience Center
University of North Carolina
115 Mason Farm Rd.,
Chapel Hill 27599-7250
Tel.: (919) 966 5807


On Fri, Apr 22, 2011 at 1:29 PM, David Knecht charter
<[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I have limited experience with multi-photon confocal microscopes.  I tried imaging live cells with one a while back and was unimpressed with the results as compared to a standard laser scanning instrument.  If you had a new single photon instrument and a new multi-photon instrument sitting side by side in your facility (and I presume some of you do), which would you choose for imaging  fixed cells or live cells in culture (not trying to go deep into tissue)?  Thanks- Dave
>
> Dr. David Knecht
> Department of Molecular and Cell Biology
> Co-head Flow Cytometry and Confocal Microscopy Facility
> U-3125
> 91 N. Eagleville Rd.
> University of Connecticut
> Storrs, CT 06269
> 860-486-2200