I second the previous lister's opinions. At the end of the day MP is really
only good for thick tissue sections. If you are looking at cell layers just
use a conventional 1P confocal. The key advantage of 2P is penetration
depth, which is moot when you are looking at cells. That said, John's
Ti:Saph beam through conventional optics than a very UV beam. This only
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Hi David,
>
> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic
> for exciting far red and infrared fluorophores. There were some exceptions,
> such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 nm
> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since you
> are ok with cost and complexity, adding an OPO(s) and/or multiple MP lasers,
> gives you full range (see
>
http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for
> OPO).
>
> Depending on cell type and physiological needs, you may be able to suppress
> phototoxicity by lowering O2 - cell culture at ~18% O2 is something of an
> artifact for many cell types - by using catalase/glucose oxidase/glucose or
> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and
> Mikhailov 1995 Cell Motil Cytokseleton).
>
> Enjoy,
>
> George
>
>
> On 4/23/2011 6:03 PM, David Knecht charter wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> Is there any disadvantage to multiphoton for cultured cells (besides cost
>> and complexity)? Cell viability? Phototoxicity? Dave
>>
>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote:
>>
>>
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>> *****
>>>
>>> Multiphoton has no clear advantage for cells in culture. For cell culture
>>> samples,
>>> we use two photon only to excite DAPI or other UV and near-UV fluors
>>> since we
>>> had to make a choice between the Ti:S and a 405 laser on our scope.
>>>
>>> Kate Luby-Phelps
>>>
>>>
>> Dr. David Knecht
>> Department of Molecular and Cell Biology
>> Co-head Flow Cytometry and Confocal Microscopy Facility
>> U-3125
>> 91 N. Eagleville Rd.
>> University of Connecticut
>> Storrs, CT 06269
>> 860-486-2200
>> 860-486-4331 (fax)
>>
>>
>>
>
>
> --
>
>
> George McNamara, PhD
> Analytical Imaging Core Facility
> University of Miami
>