the plane of focus. Seems that 2P would overcome this problem.
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> I second the previous lister's opinions. At the end of the day MP is really
> only good for thick tissue sections. If you are looking at cell layers just
> use a conventional 1P confocal. The key advantage of 2P is penetration
> depth, which is moot when you are looking at cells. That said, John's
> comments about shorter UV imaging are still valid; it's easier to get a NIR
> Ti:Saph beam through conventional optics than a very UV beam. This only
> applies if you are using dyes that need very UV excitation of course.
>
> Craig
>
>
>
>
> On Sun, Apr 24, 2011 at 6:29 AM, George McNamara
> <
[hidden email]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> Hi David,
>>
>> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic
>> for exciting far red and infrared fluorophores. There were some exceptions,
>> such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 nm
>> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since you
>> are ok with cost and complexity, adding an OPO(s) and/or multiple MP lasers,
>> gives you full range (see
>>
http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for
>> OPO).
>>
>> Depending on cell type and physiological needs, you may be able to suppress
>> phototoxicity by lowering O2 - cell culture at ~18% O2 is something of an
>> artifact for many cell types - by using catalase/glucose oxidase/glucose or
>> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and
>> Mikhailov 1995 Cell Motil Cytokseleton).
>>
>> Enjoy,
>>
>> George
>>
>>
>> On 4/23/2011 6:03 PM, David Knecht charter wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>> *****
>>>
>>> Is there any disadvantage to multiphoton for cultured cells (besides cost
>>> and complexity)? Cell viability? Phototoxicity? Dave
>>>
>>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote:
>>>
>>>
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>> *****
>>>>
>>>> Multiphoton has no clear advantage for cells in culture. For cell culture
>>>> samples,
>>>> we use two photon only to excite DAPI or other UV and near-UV fluors
>>>> since we
>>>> had to make a choice between the Ti:S and a 405 laser on our scope.
>>>>
>>>> Kate Luby-Phelps
>>>>
>>>>
>>> Dr. David Knecht
>>> Department of Molecular and Cell Biology
>>> Co-head Flow Cytometry and Confocal Microscopy Facility
>>> U-3125
>>> 91 N. Eagleville Rd.
>>> University of Connecticut
>>> Storrs, CT 06269
>>> 860-486-2200
>>> 860-486-4331 (fax)
>>>
>>>
>>>
>>
>>
>> --
>>
>>
>> George McNamara, PhD
>> Analytical Imaging Core Facility
>> University of Miami
>>