no threshold. So, the uncaging volume at low powers is the same as
the PSF and for 2P its the 2P PSF. If ground state depletion develops
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> How much uncaging occurs outside the focal region in single photon?
> If you
> don't have your laser turned up to much the energy density should only
> exceed the uncaging threshold near the focal point, yes?
>
> Craig
>
>
> On Wed, Apr 27, 2011 at 6:48 PM, Rosemary White <
[hidden email]
> >wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> Dear all,
>>
>> Haven't followed this discussion closely, but would one advantage
>> of 2P lie
>> in the ability to uncage caged compounds - fluorescent tracers, for
>> example,
>> within a single cell? We've used UV uncaging and while you get most
>> uncaging in the target cell, you also get cones of uncaging above
>> and below
>> the plane of focus. Seems that 2P would overcome this problem.
>>
>> Rosemary White
>>
>> Dr Rosemary White
>> CSIRO Plant Industry
>> GPO Box 1600
>> Canberra, ACT 2601
>> Australia
>>
>> T 61 2 6246 5475
>> F 61 2 6246 5334
>> E
[hidden email]
>>
>>
>>
>> On 28/04/11 10:10 AM, "Craig Brideau" <
[hidden email]>
>> wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>> *****
>>>
>>> I second the previous lister's opinions. At the end of the day MP
>>> is
>> really
>>> only good for thick tissue sections. If you are looking at cell
>>> layers
>> just
>>> use a conventional 1P confocal. The key advantage of 2P is
>>> penetration
>>> depth, which is moot when you are looking at cells. That said,
>>> John's
>>> comments about shorter UV imaging are still valid; it's easier to
>>> get a
>> NIR
>>> Ti:Saph beam through conventional optics than a very UV beam.
>>> This only
>>> applies if you are using dyes that need very UV excitation of
>>> course.
>>>
>>> Craig
>>>
>>>
>>>
>>>
>>> On Sun, Apr 24, 2011 at 6:29 AM, George McNamara
>>> <
[hidden email]>wrote:
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>> *****
>>>>
>>>> Hi David,
>>>>
>>>> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be
>>>> problematic
>>>> for exciting far red and infrared fluorophores. There were some
>> exceptions,
>>>> such as Alexa Fluor 488m 568, 594, and 633 all exciting well
>>>> around 755
>> nm
>>>> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but
>>>> since
>> you
>>>> are ok with cost and complexity, adding an OPO(s) and/or multiple
>>>> MP
>> lasers,
>>>> gives you full range (see
>>>>
http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range
>>>> for
>>>> OPO).
>>>>
>>>> Depending on cell type and physiological needs, you may be able to
>> suppress
>>>> phototoxicity by lowering O2 - cell culture at ~18% O2 is
>>>> something of
>> an
>>>> artifact for many cell types - by using catalase/glucose oxidase/
>>>> glucose
>> or
>>>> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer
>>>> 1993 and
>>>> Mikhailov 1995 Cell Motil Cytokseleton).
>>>>
>>>> Enjoy,
>>>>
>>>> George
>>>>
>>>>
>>>> On 4/23/2011 6:03 PM, David Knecht charter wrote:
>>>>
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>>> *****
>>>>>
>>>>> Is there any disadvantage to multiphoton for cultured cells
>>>>> (besides
>> cost
>>>>> and complexity)? Cell viability? Phototoxicity? Dave
>>>>>
>>>>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote:
>>>>>
>>>>>
>>>>>
>>>>>> *****
>>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>>>>>> *****
>>>>>>
>>>>>> Multiphoton has no clear advantage for cells in culture. For cell
>> culture
>>>>>> samples,
>>>>>> we use two photon only to excite DAPI or other UV and near-UV
>>>>>> fluors
>>>>>> since we
>>>>>> had to make a choice between the Ti:S and a 405 laser on our
>>>>>> scope.
>>>>>>
>>>>>> Kate Luby-Phelps
>>>>>>
>>>>>>
>>>>> Dr. David Knecht
>>>>> Department of Molecular and Cell Biology
>>>>> Co-head Flow Cytometry and Confocal Microscopy Facility
>>>>> U-3125
>>>>> 91 N. Eagleville Rd.
>>>>> University of Connecticut
>>>>> Storrs, CT 06269
>>>>> 860-486-2200
>>>>> 860-486-4331 (fax)
>>>>>
>>>>>
>>>>>
>>>>
>>>>
>>>> --
>>>>
>>>>
>>>> George McNamara, PhD
>>>> Analytical Imaging Core Facility
>>>> University of Miami
>>>>
>>