> The primary dichroic is typically the largest problem. Our Nikon
> technician went through some fairly heroic efforts to install a dichroic
> into an A1 FN1. We had the filter custom made at significant expense for
> our ultrabroadband Ti:Saph laser (Octavius), so the possibility of damaging
> the filter during installation was a critical concern. After the technician
> partly disassembled the scan head in our lab, we ended up drafting a nice
> lady with very tiny fingers who did needlepoint and crocheting as a hobby to
> finally get the filter into place in the A1 scan head. This was something
> normally done at the factory... That said, it did work out for us and Nikon
> was very accommodating in helping us with the modification.
> Regarding piping the laser in to the Zeiss scope: I was assuming your laser
> could be brought in free space. If you have to go through a fiber then you
> probably won't be able to do it due to dispersion in the glass fiber. If
> you have a prechirping unit then you *might* be able to pull it off, but it
> will be tricky.
>
> Craig
>
>
>
> On Mon, May 9, 2011 at 4:06 PM, Armstrong, Brian <
[hidden email]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> Craig makes a good point. The HFT700/488 or KP650 (for example) that you
>> have been using for your 2P laser will not work for a 405 line. Therefore
>> you would need to change the dichroics. This is not done in the field and
>> you would need to send your scan-head to the factory in Germany. You would
>> also need to think about how to get your laser into the fiber (AOTF?), how
>> to adjust the power (AOM is no longer useful), and whether or not you have
>> the secondary dichroics that will pass the emitted light.
>>
>> You should talk to Zeiss before considering this idea any further.
>>
>> Cheers,
>>
>>
>> Brian D Armstrong PhD
>> Assistant Research Professor
>> Light Microscopy Core
>> Beckman Research Institute
>> City of Hope
>> Dept of Neuroscience
>> 1450 E Duarte Rd
>> Duarte, CA 91010
>> 626-256-4673 x62872
>>
>>
>>
http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx>>
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:
[hidden email]]
>> On Behalf Of Craig Brideau
>> Sent: Monday, May 09, 2011 2:52 PM
>> To:
[hidden email]
>> Subject: Re: 405 nm laser via NIR port on Zeiss 510
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> For starters, what do you have for a Dichroic that could pipe in the 405
>> and
>> let the rest of the visible pass? Most 2p dichroics split off at around
>> 680
>> or 700nm to allow the IR to pass and the emission light to reflect, or
>> vice-versa.
>>
>> Craig
>>
>>
>> On Mon, May 9, 2011 at 1:16 PM, Phillips, Thomas E.
>> <
[hidden email]>wrote:
>>
>> > *****
>> > To join, leave or search the confocal microscopy listserv, go to:
>> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> > *****
>> >
>> > The NIR laser for two photon excitation on our Zeiss 510 died on us. The
>> > repair would be at least 60K. We are trying to find funds for a new
>> confocal
>> > so I don't know if it is worth repairing. The biggest practical problem
>> is
>> > that we can't do DAPI excitation on our routine preps. Has anyone tried
>> > retrofitting a 405 nm laser through the NIR port? Are there optical
>> > components/mirrors in the pathway that only work with longer
>> wavelengths?
>> > Any thoughts on this are welcome. Tom
>> >
>> >
>> > Thomas E. Phillips, Ph.D
>> > Professor of Biological Sciences
>> > Director, Molecular Cytology Core
>> > 2 Tucker Hall
>> > University of Missouri
>> > Columbia, MO 65211-7400
>> > 573-882-4712 (office)
>> > 573-882-0123 (fax)
>> >
[hidden email]<mailto:
[hidden email]>
>> >
>> >
http://www.biology.missouri.edu/faculty/phillips.html>> >
http://www.biotech.missouri.edu/mcc/>> >
>>
>>
>> ---------------------------------------------------------------------
>> SECURITY/CONFIDENTIALITY WARNING:
>> This message and any attachments are intended solely for the individual or
>> entity to which they are addressed. This communication may contain
>> information that is privileged, confidential, or exempt from disclosure
>> under applicable law (e.g., personal health information, research data,
>> financial information). Because this e-mail has been sent without
>> encryption, individuals other than the intended recipient may be able to
>> view the information, forward it to others or tamper with the information
>> without the knowledge or consent of the sender. If you are not the intended
>> recipient, or the employee or person responsible for delivering the message
>> to the intended recipient, any dissemination, distribution or copying of the
>> communication is strictly prohibited. If you received the communication in
>> error, please notify the sender immediately by replying to this message and
>> deleting the message and any accompanying files from your system. If, due to
>> the security risks, you do not wish to receive further communications via
>> e-mail, please reply to this message and inform the sender that you do not
>> wish to receive further e-mail from the sender.
>>
>> ---------------------------------------------------------------------
>>
>
>