I would recommend looking at the dish membrane without any fluorophore labeling
to check for autofluorescence. I've had a couple people use membranes that are
autofluorescent - mostly in the 488nm range.
So red or far red phalloidin would be better in this case, perhaps.
- Lisa


---------------------------------------
Lisa Cameron, Ph.D.
Director of Confocal and Light Microscopy Core
Dana Farber Cancer Institute
Boston, MA 02215
[hidden email]


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Renato Mortara
Sent: Tuesday, June 14, 2011 10:30 AM
To: [hidden email]
Subject: [CONFOCALMICROSCOPY] Phalloidin background on transwell membrane

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear all
 
A colleague of mine is trying to label polarized epithelial cells grown on
transwell dishes.
 
However, fluorescent phalloidin is labeling the dish membrane, rendering it
impossible to image the cells.
 
She tried to block with goat serum, BSA with no luck.
 
Any suggestions ?
 
Many thanks !
 
Renato
 
Renato A. Mortara
Parasitology Division
UNIFESP - Escola Paulista de Medicina
Rua Botucatu, 862, 6th floor
São Paulo, SP
04023-062
Brazil
Phone: 55 11 5579-8306
Fax:     55 11 5571-1095
email: [hidden email]
home page: www.ecb.epm.br/~ramortara


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.