Re: averaging vs. accumulation for noise reduction - is there a difference?

Posted by Guy Cox-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/averaging-vs-accumulation-for-noise-reduction-is-there-a-difference-tp6483751p6483832.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Mark,

    By my reckoning it's after 4 in the morning in Auckland - what are
you doing replying at this hour?  (By the way, I'm in Taiwan, not Oz).

                                   Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net
 


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Mark Cannell
Sent: Friday, 17 June 2011 2:41 AM
To: [hidden email]
Subject: Re: averaging vs. accumulation for noise reduction - is there a
difference?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

It also depends on how the readout  bandwidth is controlled for
different scan speeds...

Cheers
On 16/06/2011, at 5:26 PM, Stanislav Vitha wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hallo,
> this is a very basic question, but I cannot figure this out from what
I have
> been reading, so a simple explanation for a non-physicist would be
much
> appreciated:
>
> Is there a real difference in the improvement of the signal to noise
ratio
> between frame averaging (or accumulation) and longer dwell times
(slower
> scan) for a point-scanning confocal witrh a PMT detector?
>
> For instance, using single point scanning confocal, 12-bit
acquisition.
>
> a) averaging (or accumulating) 5 frames, 4 microseconds per pixel
> b) acquiring a single frame, 20 microseconds per pixel
>
> Assumptions:
> no saturation of the detector;
> stable environmental conditions, no focus drift, etc
>
> Would it matter (for the dfference between the two scenarios) if it
was analog
> detection or photon counting detection?
>
> I will run this little test later, but I am curious what you think.
>
> I thought that at least for the photon counting mode, the two
important
> factors would be the dark counts and the total number of counts
detected, so
> whether it is acquired in one scan or in 5 scans, it should be the
same. My
> camera expert here insists that the averaging scheme will give better
noise
> suppression.
>
> Thanks!
>
>
> Stan Vitha  
>
> Microscopy and Imaging Center
> Texas A&M University

-----