I have not done a careful analysis of this either, and I am not quite sure how you would, however I share the viewpoint of Julio exactly, 1.6usec/pix and 2-4 ave.
Brian Armstrong PhD
Light Microscopy Core
Beckman Research Institute
1450 East Duarte Rd
Duarte, CA 91010
626-256-4673 x62872
http://www.cityofhope.org/SharedResources/LightMicroscopy/LightMicroHome.htm
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Julio Vazquez
Sent: Thursday, June 16, 2011 2:21 PM
To:
[hidden email]
Subject: Re: averaging vs. accumulation for noise reduction - is there a difference?
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
This is what I noticed empirically on our Zeiss LSM 510, where averaging tends to give somewhat better noise reduction than increasing dwell time. Under "normal" imaging conditions, we typically use a dwell time of 1.6-3.2 microseconds. Increasing the dwell time to greater than 3.2 microseconds tends to result in more bleaching and somewhat reduced signal. Typically, we use 1.6 microseconds dwell time, and 2-4 averages, depending on the sample.
--
Julio Vazquez
Fred Hutchinson Cancer Research Center
Seattle, WA
http://www.fhcrc.org/On Jun 16, 2011, at 2:08 PM, Moninger, Thomas wrote:
> Stan,
>
> I've been told by Carl Z. engineers that in general averaging (I usually use line, not frame) tends to yield better S/N then does increasing dwell time. As Lloyd commented this may be model specific. I have not done any analysis to confirm this however....
>
> Tom
---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING:
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.
---------------------------------------------------------------------