Re: Incubator box heating mystery

Posted by Craig Brideau on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Incubator-box-heating-mystery-tp6627357p6631050.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

We had quite a bit of trouble with our room ventilation until quite
recently. We purchased an inexpensive USB datalogger to continuously record
temperature and humidity in the room.  It picked up large swings in
temperature and humidity that the ventilation system was supposed to
prevent.  The data we collected finally convinced the building contractor to
install proper humidity controls and hardware for our room.  I recommend
that every facility should keep logs of temperature and humidity.  The
loggers are quite cheap now, and can be programmed, detached from the
computer, and left unattended for months.  We would move ours around the
room to get a feel for any gradients, heat loads from equipment, etc.

Craig

On Thu, Jul 28, 2011 at 11:03 AM, Zac Arrac Atelaz <[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
>
> Esteban:
>
> Your problem might be the changing position of the termometer of the
> system, one interesting trial will be putting the overheating system in the
> non overheating system, to see if is that part fails or remain working as
> the trial you make the way around.
>
> If your system is failing you should consider one incubator as the one we
> have, this is not the huge box making the microscope unreachable, it is the
> size of the insert in the stage and it has 4 points heating your sample, the
> cover, the objective used, and the water heater, we have never had a
> temperature overshoot as the mentioned by Michael, even if we open doors, or
> change room temperature ( recorded from 18 to 27°C) the shift in temperature
> observed trough time in the incubator is about 0.5°C over a 25h period. In
> such long experiments we have people reviewing samples as you dont want to
> let the confocal working if the sample has suffered unwanted changes, by the
> way the brand of the one we have is a INU - Tokai hit incubator.
>
> I hope this helps a little.
>
> Gabriel Orozco Hoyuela
>
>
> > Date: Wed, 27 Jul 2011 12:01:32 -0700
> > From: [hidden email]
> > Subject: Incubator box heating mystery
> > To: [hidden email]
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Hello everyone,
> >
> > I have a very strage problem with an incubator box heating up. I have two
> > live cell imaging microscopes enclosed in PeCon incubator boxes (I think
> > they are Incubator XL). The microscopes are in two separate buildings.
> One
> > microscope heats up significantly past the set temp. (above 41 degrees,
> set
> > to 37 degrees) even though the temp. is being measured and reported
> > correctly to the electronics (confirmed by a glass thermometer). The
> other
> > microscope works perfectly and holds at 37 degrees for days. The strange
> > thing is that when I put the temp. control components that work well
> > (heater, control electronics, temp. sensor, and all cables) onto the
> > microscope that heats up, it still heats up, even though the components
> work
> > perfectly on the other microscope! Any ideas why this might be happening,
> > why the temp. control equipment works on one microscope but not on the
> > other?
> >
> > The temp. equipment is stand-alone, not connected to a computer or to the
> > microscope in any way. One microscope is a Zeiss Axiovert 200M (the one
> > that works) and the other is a Leica DM IRE2 (heats up) in separate
> > buildings; the incubator boxes surrounding them are very similar. Any
> ideas
> > on this mystery are welcome.
> >
> > Thanks,
> > Esteban
>
>