Re: Incubator box heating mystery

Posted by Craig Brideau on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Incubator-box-heating-mystery-tp6627357p6632108.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

It's hard to get, but really you want +/-1 degree C and humidity <35%.  This
is particularly important if you have a lot of free-space optics.  Most
fiber-coupled systems can tolerate some temperature change, but free-space
beams are guided by steering mirrors.  Temperature changes can cause the
mirror mounts to flex which misaligns your beam.  Additionally, Ti:Saph
lasers used in two-photon HATE humidity.  We're in Calgary, Canada, which is
very, very dry in the winter.  We can run our lasers without a dry air or
nitrogen purge at 925nm.  In the summer though, the humidity gets quite high
and purging becomes necessary.  The building contractor just finished
installing a humidity control system for our lab last week, so we will see
how things go.  Thus far it seems to be holding the humidity under about
37%.  I'll wait a couple weeks and then check my datalogger.  @:-)

Craig


On Thu, Jul 28, 2011 at 6:59 PM, Graham Wright <
[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I agree with Craig that the temperature and humidity loggers give you
> a real insight into what's going on with the aircon in your rooms
> (particularly over the weekends in our case). I'd be interested though
> in what you think are reasonable limits for the tolerable fluctuations
> in both temperature and humidity. We are constantly battling with both
> in Singapore!
>
> Regards,
> Graham
>
> ---
> Graham Wright, PhD
> Microscopy Unit Manager
>
> Institute of Medical Biology
> 8A Biomedical Grove, #06-06 Immunos, Singapore 138648
>
> E:   [hidden email]
> W:  http://www.imb.a-star.edu.sg/imu/
>
>
>
> On 29 July 2011 02:51, Craig Brideau <[hidden email]> wrote:
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > We had quite a bit of trouble with our room ventilation until quite
> > recently. We purchased an inexpensive USB datalogger to continuously
> record
> > temperature and humidity in the room.  It picked up large swings in
> > temperature and humidity that the ventilation system was supposed to
> > prevent.  The data we collected finally convinced the building contractor
> to
> > install proper humidity controls and hardware for our room.  I recommend
> > that every facility should keep logs of temperature and humidity.  The
> > loggers are quite cheap now, and can be programmed, detached from the
> > computer, and left unattended for months.  We would move ours around the
> > room to get a feel for any gradients, heat loads from equipment, etc.
> >
> > Craig
> >
> > On Thu, Jul 28, 2011 at 11:03 AM, Zac Arrac Atelaz <[hidden email]
> >wrote:
> >
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> *****
> >>
> >>
> >> Esteban:
> >>
> >> Your problem might be the changing position of the termometer of the
> >> system, one interesting trial will be putting the overheating system in
> the
> >> non overheating system, to see if is that part fails or remain working
> as
> >> the trial you make the way around.
> >>
> >> If your system is failing you should consider one incubator as the one
> we
> >> have, this is not the huge box making the microscope unreachable, it is
> the
> >> size of the insert in the stage and it has 4 points heating your sample,
> the
> >> cover, the objective used, and the water heater, we have never had a
> >> temperature overshoot as the mentioned by Michael, even if we open
> doors, or
> >> change room temperature ( recorded from 18 to 27°C) the shift in
> temperature
> >> observed trough time in the incubator is about 0.5°C over a 25h period.
> In
> >> such long experiments we have people reviewing samples as you dont want
> to
> >> let the confocal working if the sample has suffered unwanted changes, by
> the
> >> way the brand of the one we have is a INU - Tokai hit incubator.
> >>
> >> I hope this helps a little.
> >>
> >> Gabriel Orozco Hoyuela
> >>
> >>
> >> > Date: Wed, 27 Jul 2011 12:01:32 -0700
> >> > From: [hidden email]
> >> > Subject: Incubator box heating mystery
> >> > To: [hidden email]
> >> >
> >> > *****
> >> > To join, leave or search the confocal microscopy listserv, go to:
> >> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> > *****
> >> >
> >> > Hello everyone,
> >> >
> >> > I have a very strage problem with an incubator box heating up. I have
> two
> >> > live cell imaging microscopes enclosed in PeCon incubator boxes (I
> think
> >> > they are Incubator XL). The microscopes are in two separate buildings.
> >> One
> >> > microscope heats up significantly past the set temp. (above 41
> degrees,
> >> set
> >> > to 37 degrees) even though the temp. is being measured and reported
> >> > correctly to the electronics (confirmed by a glass thermometer). The
> >> other
> >> > microscope works perfectly and holds at 37 degrees for days. The
> strange
> >> > thing is that when I put the temp. control components that work well
> >> > (heater, control electronics, temp. sensor, and all cables) onto the
> >> > microscope that heats up, it still heats up, even though the
> components
> >> work
> >> > perfectly on the other microscope! Any ideas why this might be
> happening,
> >> > why the temp. control equipment works on one microscope but not on the
> >> > other?
> >> >
> >> > The temp. equipment is stand-alone, not connected to a computer or to
> the
> >> > microscope in any way. One microscope is a Zeiss Axiovert 200M (the
> one
> >> > that works) and the other is a Leica DM IRE2 (heats up) in separate
> >> > buildings; the incubator boxes surrounding them are very similar. Any
> >> ideas
> >> > on this mystery are welcome.
> >> >
> >> > Thanks,
> >> > Esteban
> >>
> >>
> >
>