Re: Uniformity of 2-photon illumination?

Posted by Guy Cox-2 on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Uniformity-of-2-photon-illumination-tp6719119p6719167.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

The Chroma slide should be a pretty good test.  You should do 2 or 3
focal positions in case the effect is actually a tilt in the stage or
the optics.  Do you see the same effect in the confocal and
non-descanned detectors?  If so it's clearly not something interfering
with the detection optics, and you'd have to blame the illumination.  

                                              Guy

Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006

Phone +61 2 9351 3176     Fax +61 2 9351 7682
             Mobile 0413 281 861
______________________________________________
      http://www.guycox.net
 


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of stu_the_flat
Sent: Wednesday, 24 August 2011 5:14 PM
To: [hidden email]
Subject: Uniformity of 2-photon illumination?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear list users,

We have a Zeiss 510 two-photon system, it seems to have a an non-uniform
field of illumination,

I have tried to test this by imaging a chroma fluorescent test slide
there
is a two fold drop in fluorescent signal across the X axis, the there
appears to be a slight variation on the Y axis as well,

What is most likely to be causing this? Is imaging a test slide a fair
test?

Any advice would be most welcome.

Thank you

Stuart McIntyre


--
View this message in context:
http://confocal-microscopy-list.588098.n2.nabble.com/Uniformity-of-2-pho
ton-illumination-tp6719119p6719119.html
Sent from the Confocal Microscopy List mailing list archive at
Nabble.com.

-----
No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1392 / Virus Database: 1520/3853 - Release Date: 08/23/11