Re: Uniformity of 2-photon illumination?

Posted by Armstrong, Brian on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Uniformity-of-2-photon-illumination-tp6719119p6724897.html

Hi Stuart, in short, I align the 2P beam just as described by Julio in previous e-mail reply. I suggest you call your Zeiss service rep and then watch how he aligns the 2P to the vis laser(s). Once you've seen it, the protocol should be very easy to replicate.
You may already have a saved configuration called "2P Align" in C:AIM/settings that you can use. Also, you should have received a mirror slide with a grid on it that is perfect for this kind of alignment.

Adjusting the beam path as Julio described is quite easy to do.
Also, don't look into the laser light with your remaining good eye (first heard this sound advice from George McNamara).

Cheers,  

Brian D Armstrong PhD
Assistant Research Professor
Light Microscopy Core
Beckman Research Institute
City of Hope
Dept of Neuroscience
1450 E Duarte Rd
Duarte, CA 91010
626-256-4673 x62872

http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of stu_the_flat
Sent: Thursday, August 25, 2011 3:38 AM
To: [hidden email]
Subject: Re: Uniformity of 2-photon illumination?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Thank you very much for your replies.

 

Just to clear up a few points, I always
take Z sections of the slide just to make sure it wasn’t a case that the slide
was at an angle, (The system is not designed to hold slides so its always a bit
“Heath Robison” trying to get a slide underneath it!)

 

I had checked various sites around the
slide and I was getting consistent results, yesterday I scanned the slide with
the descanned detectors and got a similar image.

 

We want all the power we can get as would
love to be able to image more deeply into cardiac muscle, (is it a fair
assumption that more power will allow us to image deeper?) for that reason we
have a Coherent Chameleon Ultra II, and I believe its peak power is around
5.5W. However would the alignment not take place “down stream” of the OPO? So I
could just set that to its minimum intensity setting.

 

Anyway it is kind of irrelevant because the
system is under service contract with Zeiss, So all I need to do it get on the
phone and start whining at them, Although I would love to give laser alignment a
go myself!

 

Once again thank you for all your replies

 

Yours

Stuart McIntyre



--
View this message in context: http://confocal-microscopy-list.588098.n2.nabble.com/Uniformity-of-2-photon-illumination-tp6719119p6723940.html
Sent from the Confocal Microscopy List mailing list archive at Nabble.com.


---------------------------------------------------------------------
*SECURITY/CONFIDENTIALITY WARNING:  
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.

---------------------------------------------------------------------