Or the other way round: Uneven illumination is a perfect sign that the beam path is not ok (but a service technician should be able to fix it).
> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:
[hidden email]] On Behalf Of Hanna Sas
> Nowosielska
> Sent: jeudi 6 octobre 2011 16:37
> To:
[hidden email]
> Subject: Re: Rejected posting to
[hidden email]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Yes, at least Olymus claims everything with the optical path is ok.
>
> Hanna
>
> 2011/10/6 tineke vendrig <
[hidden email]>:
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > *****
> >
> > Have you optimised you microscope ?? Kohler illumination and so on..??
> >
> > Tineke Vendrig, ing
> > technical engeneer optical microscopy
> > Delft University of Technology
> > Bionano Science
> > Kavli Institute of Nanoscience
> > Lorentzweg 1
> > 2628LJ Delft
> > room F185
> > Tel: +31 27 89299
> > Fax:+31 15 2781202
> > email:
[hidden email]
> > mobile phone: 06-34241412
> >
> >
> > 2011/10/6 Hanna Sas Nowosielska <
[hidden email]>
> >
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> >> *****
> >>
> >> Dear all,
> >>
> >> For a couple of months we are struggling with a serious problem with
> >> the field illumination in our Olympus Scan^R Scanning station. The
> >> image of the specimen, when observed on the computer screen is much
> >> darker on one side than on the other. We thought it is the lamp but
> >> the lamp is new at properly inserted. Then we tried calibrating the
> >> optic fiber, but it didn't help at all. Our third guess was the
> >> filters.We checked all of the filters - the coating has lots of
> >> small dark spots with biggest concentration of the spots in the
> >> middle of the filters, but I am concerned if that can produce uneven
> >> illumination of the field so that only one part (let say left corner)
> >> is underilluminated? So, I turn to you and your experience: did may
> >> anyone have the similar problem? Do you have any suggestion what can
> >> be the cause of the problem?
> >>
> >> Best regards,
> >> Hanna
> >>
> >> --
> >> Department of Plant Anatomy and Cytology Faculty of Biology and
> >> Environmental Protection University of Silesia Jagiellonska Str. 28
> >> 40-032 Katowice
> >> Poland
> >>
> >
> >
> >
> > --
> > Tineke Vendrig, ing
> > technical engeneer optical microscopy
> > Delft University of Technology
> > Bionano Science
> > Kavli Institute of Nanoscience
> > Lorentzweg 1
> > 2628LJ Delft
> > room F185
> > Tel: +31 27 89299
> > Fax:+31 15 2781202
> > email:
[hidden email]
> > mobile phone: 06-34241412
> >
>
>
>
> --
> Department of Plant Anatomy and Cytology Faculty of Biology and
> Environmental Protection University of Silesia Jagiellonska Str. 28
> 40-032 Katowice
> Poland