Oops--I accidentally inverted the equation. It should be QY=QYstandard*(F/Fstandard) *(Astandard/A)*(n^2/nstandard^2).
For quantum yield standards, see Magde et al photochemistry and photobiology 2002 vol 75 p 327. It gives a thorough study for fluorescein and rhodamine 6G. Other wavelengths are a bit more difficult. Lakowicz principles of fluorescence spectroscopy gives more details and a few standards from the literature.
That's a pretty neat method! It avoids having to worry about figuring out photon flux by calibrating against a known. Where would you get trustworthy QE calibration standards though?
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> In order to measure quantum yields by the reference method, one only
> needs a defined measurement volume, a standard dye of known absorbance
> (at the microscope excitation wavelength) and quantum yield, and the
> absorbance of your sample at the microscope excitation wavelength.
> Quantum yield is then defined as QY=QYstandard*(A/Astandard) *(Fstandard/F)*(n^2/nstandard^2).
> Here n is the refractive index, F is the measured fluorescence
> intensity and A is the absorbance at the microscope excitation
> wavelength. Note that it is important that the standard and your
> unknown have similar emission spectra. Otherwise you will have to
> correct for the wavelength dependence of your microscope detection
> efficiency. Of course, if you have enough sample to measure
> absorbance, you can also measure fluorescence in a cuvette and then there is no point in going to the microscope.
>
> Jay
>
> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:
[hidden email]]
> On Behalf Of Guy Cox
> Sent: Thursday, October 06, 2011 8:44 PM
> To:
[hidden email]
> Subject: Quantum yield
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> I got a rather left field enquiry today, as to whether there were
> calibration standards for quantum yield. It seems that the person
> wants to measure quantum yield under the microscope. My immediate
> response was that this is impossible. Quantum yield is easy enough to
> measure in a cuvette but would it be possible in a microscope? You
> could make a standard of a known concentration of fluorescein in a cell made by a
> spacer under the coverslip, but where do you go from there, if both
> quantum yield and extinction coefficient of the test sample are unknown?
>
> Any bright ideas?
>
> Guy
>
>
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox CRC Press / Taylor & Francis
>
http://www.guycox.com/optical.htm> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for
> Microscopy & Microanalysis, Madsen Building F09, University of Sydney,
> NSW
> 2006
>
> Phone +61 2 9351 3176 Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>
http://www.guycox.net>
>