Dear Hanna
Koehler comes to mind as already suggested.
I m not familiar with scanning station
We had similar problem on an Olympus scope and found that the glass in front of the transmission lamp had a crack (heat?)
So, this glass in the transmission housing, which directly faces the microscope stage , had a crack and that produced a shadow in the image pretending to be a Koehler problem
cheers
Axel 6-19B , 92715622
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Hanna Sas Nowosielska
Sent: Thursday, 6 October, 2011 10:37 PM
To:
[hidden email]
Subject: Re: Rejected posting to
[hidden email]
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Yes, at least Olymus claims everything with the optical path is ok.
Hanna
2011/10/6 tineke vendrig <
[hidden email]>:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Have you optimised you microscope ?? Kohler illumination and so on..??
>
> Tineke Vendrig, ing
> technical engeneer optical microscopy
> Delft University of Technology
> Bionano Science
> Kavli Institute of Nanoscience
> Lorentzweg 1
> 2628LJ Delft
> room F185
> Tel: +31 27 89299
> Fax:+31 15 2781202
> email:
[hidden email]
> mobile phone: 06-34241412
>
>
> 2011/10/6 Hanna Sas Nowosielska <
[hidden email]>
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> Dear all,
>>
>> For a couple of months we are struggling with a serious problem with
>> the field illumination in our Olympus Scan^R Scanning station. The
>> image of the specimen, when observed on the computer screen is much
>> darker on one side than on the other. We thought it is the lamp but
>> the lamp is new at properly inserted. Then we tried calibrating the
>> optic fiber, but it didn't help at all. Our third guess was the
>> filters.We checked all of the filters - the coating has lots of
>> small dark spots with biggest concentration of the spots in the
>> middle of the filters, but I am concerned if that can produce uneven
>> illumination of the field so that only one part (let say left corner)
>> is underilluminated? So, I turn to you and your experience: did may
>> anyone have the similar problem? Do you have any suggestion what can
>> be the cause of the problem?
>>
>> Best regards,
>> Hanna
>>
>> --
>> Department of Plant Anatomy and Cytology Faculty of Biology and
>> Environmental Protection University of Silesia Jagiellonska Str. 28
>> 40-032 Katowice
>> Poland
>>
>
>
>
> --
> Tineke Vendrig, ing
> technical engeneer optical microscopy
> Delft University of Technology
> Bionano Science
> Kavli Institute of Nanoscience
> Lorentzweg 1
> 2628LJ Delft
> room F185
> Tel: +31 27 89299
> Fax:+31 15 2781202
> email:
[hidden email]
> mobile phone: 06-34241412
>
--
Department of Plant Anatomy and Cytology Faculty of Biology and Environmental Protection University of Silesia Jagiellonska Str. 28
40-032 Katowice
Poland
Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.