Every objective needs to be tested with beads with the wavelengths you are using.  Any time we need precise XY or Z axis registration we test with beads or fluorescent coatings on the surfaces in the same media as the biological imaging, sometimes even on the cells.  Years ago we found out the hard way, when doing CFP/YFP ratio FRET, that there is no substitute for characterizing each objective as close to the real imaging conditions as practical.  Generalizations such as, "High NA brand X planapo objective are fine," just are not true.
_________________________________________
Michael Cammer, Assistant Research Scientist
Skirball Institute of Biomolecular Medicine
Lab: (212) 263-3208  Cell: (914) 309-3270

------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================