A few weeks ago I posted a message that we were having strange fluctuations of intensity with our Zeiss 710 confocal. Turns out the AOTF controller was shot. Also, there was a question whether the "crystal" had gone bad. Regardless, Zeiss service quickly fixed the problem and tuned up the system with more laser power than before it went down and it has been stable since (about 6 weeks so far).
Regards,
Michael
________________________________________________________
Michael Cammer, Assistant Research Scientist
Skirball Institute of Biomolecular Medicine
Lab: (212) 263-3208 Cell: (914) 309-3270
-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of Jeremy Adler
Sent: Thursday, February 09, 2012 3:25 AM
To:
[hidden email]
Subject: Re: laser stability
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
If you cannot prevent the fluctuations in laser power you can attempt
to correct for them.
Zeiss used to have a diode to detect laser power (our one never worked
properly), which will record an image that records the laser power -
use this image to correct ( by division) the fluorescence images.
Similarly, depending in the sample, you can collect a transmitted
light image while collecting the fluorescence. This will record the
laser power and can also be used as a correction.
For the correction the laser power image and the fluorescent image
need to have zero intensity set correctly. Then divide the fluorescent
image with the laser intensity and rescale by the mean laser
intensity. Not perfect and assumes a linear relationship between power
and fluorescence, but will make a big improvement.
Quoting Arvydas Matiukas <
[hidden email]>:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Dear list,
>
> Please advice what is the most efficient/cheap way to deal
> with confocal laser instability. Our LSM 510 is 7 years old
> but Ar and Hene 543 lasers have been replaced two years
> ago and have been used no more that 2000-2500 hrs.
> Users started complaining about 488nm and 543nm laser
> line stability, and I measured it using Chroma fluorescent
> plastic slides and 10x objective. The frame- averaged fluorescent
> signal is reduced by 40-50% in 1hr after laser start. The decline
> curve is quite steady after first 3-5 min with some fluctuations.
>
> The most disturbing are intermittent large scale signal
> fluctuations (2-4 fold)
> occurring on about a minute timescale. They kill dynamic (e.g. FRET) or
> long-term imaging. They happen irregularly, on average once in 4-5 days,
> and always after 2-3 hrs after the lasers are started.
>
> In a related post several years ago laser cooling, laser
> polarization, AOTF cooling,
> and AOTF driver stability were listed as the key factors. I am quite
> confident that cooling
> is not an issue as the fans within the equipment boxes are working
> well, and whole room
> has a powerful local conditioning system which maintains ambient
> temperature very stable.
> Zeiss engineer checked the fibers recently, and suggested that maybe
> AOTF drivers are
> not stable. As we are not on service contract and have to pay for
> every replacement I
> would be interested to do more testing myself to identify the
> failing component.
>
> Any suggestions/advices are very welcome. I can provide the
> measurement curves if
> somebody is intersted (contact me off list).
>
> Thanks,
> Arvydas
>
>
>
>
>
>
> Arvydas Matiukas, Ph.D.
> Director of Confocal&Two-Photon Core
> Department of Pharmacology
> SUNY Upstate Medical University
> 766 Irving Ave., WH 3167
> Syracuse, NY 13210
> tel.: 315-464-7997
> fax: 315-464-8014
> email:
[hidden email]
>
Jeremy Adler
IGP
Rudbeckslaboratoriet
Daghammersköljdsväg 20
751 85 Uppsala
Sweden
0046 (0)18 471 4607
------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================