Posted by
Monique Vasseur on
URL: http://confocal-microscopy-list.275.s1.nabble.com/Deconvolution-of-Transmission-Images-tp7393699p7456471.html
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Hi John,
Indirectly, do you suggest the same for Z sampling if we are interested in 3D measurements? Thanks
Monique Vasseur
-----Message d'origine-----
De : Confocal Microscopy List [mailto:
[hidden email]] De la part de Lemasters, John J.
Envoyé : 11 avril 2012 09:34
À :
[hidden email]
Objet : Re: Nyquist and Image size
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Please remember that pixel spacing on the diagonal is 1.4 that in the horizontal and vertical directions. Accordingly to meet the Nyquist criterion for the diagonal, pixel size should be 2.3 x 1.4 = 3.2. Also, the Nyquist criterion is an arbitrary threshold, and image quality will improve somewhat with sampling greater that proposed by Nyquist. Considering diagonal sampling, I suggest using a pixel size that is one fourth of the resolving limit for the most critical work.
John
--
John J. Lemasters, MD, PhD
Professor and GlaxoSmithKline Distinguished Endowed Chair Director, Center for Cell Death, Injury & Regeneration Departments of Pharmaceutical & Biomedical Sciences and Biochemistry & Molecular Biology Medical University of South Carolina
DD504 Drug Discovery Building
70 President Street, MSC 140
Charleston, SC 29425
Office: 843-876-2360
Lab: 843-876-2354
Fax: 843-876-2353
Email:
[hidden email]
http://academicdepartments.musc.edu/ccdir-----Original Message-----
From: Confocal Microscopy List [mailto:
[hidden email]] On Behalf Of John Oreopoulos
Sent: Wednesday, April 11, 2012 8:29 AM
To:
[hidden email]
Subject: Re: Nyquist and Image size
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy*****
Renato,
Whether you have 256x256, 512x512 or 2048x2048, the "optimum" Nyquist sampling rate (ie: pixel dimensions) does not change since your objective lens did not change. The quoted pixel size at 2Kx2K you mentioned (22.5 nm x 22.5 nm) means you are oversampling the image (and not gaining anything). Your image may look smoother but it contains no more information than the 512x512 image with 90x90 nm pixel sizes. Presumably the scan speed is the same between 512x512 and 2Kx2K.
You should decrease the galvometric mirror scan zoom setting to get back to an effective pixel size of 90x90 nm with 2Kx2K pixels in your image. Effectively, you will be imaging (and properly sampling) a larger field of view then. I'm not familiar with the Leica laser scanning confocals so I'm not sure if it will allow you to do this. On other systems, like the Olympus FV300 for example, you can set your image pixel dimensions (256x256, 512x512, etc.) and your scan zoom independently.
Just out of curiosity, why image 2K x 2K when you can't easily display that on a standard computer screen or present it in a published paper without downsizing? I rarely departed from 512x512 in my laser scanning days, except when I wanted to see a larger field of view.
Cheers,
John Oreopoulos
Research Assistant
Spectral Applied Research
Richmond Hill, Ontario
Canada
www.spectral.ca
On 2012-04-11, at 7:22 AM, Renato Mortara wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Dear all,
>
> Having attended the first Pawley course in Vancouver I feel highly
> embarassed to ask this, but I would really appreciate a clarification:
>
> When estimating the highest zoom users should apply to their sample in
> order to accommodate for the Nyquist theorem, I estimated the optimum
> pixel size value by dividing the lateral resolution (eg: 0.2 microns)
> by 2.3 so that the value is approxiametely 90 nm.
>
> The doubt: if the image size is increased from 512x512 (having
> adjusted the zoom to the pixel size of 90nm) to 2Kx2K, the resulting
> pixel size (displayed by the system - Leica) the pixel size decreases
> 4 fold, to 22.5 nm. Since the resolution obviously did not change but
> only the image size, what happens to Nyquist and the optimum pixel size at 2Kx2K ?
>
> Many thanks !
>
> Renato
>
> Renato A. Mortara
> Parasitology Division
> UNIFESP - Escola Paulista de Medicina
> Rua Botucatu, 862, 6th floor
> São Paulo, SP
> 04023-062
> Brazil
> Phone: 55 11 5579-8306
> Fax: 55 11 5571-1095
> email:
[hidden email]
> home page: www.ecb.epm.br/~ramortara