I've always thought of Zoom as my field-of-view control. Many microscope
this can be either zoom or crop. Some software lets you select whether it
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Just to add a bit to John's absolutely correct explanation.
>
> It's basically a question of interpretation. You set up Nyquist imaging
> at 512x512 and then selected 2048x2048, expecting to get the same
> resolution but a 4 times bigger area. Perfectly reasonable. But the Leica
> software assumed you now wanted to get 2048x2048 pixels within the same
> chosen field of view. Also a perfectly reasonable interpretation. One
> might, in a perfect world, expect the software to ask you which
> interpretation you want, but if it doesn't it's pretty easy to fix.
>
> Guy
>
>
> Optical Imaging Techniques in Cell Biology
> by Guy Cox CRC Press / Taylor & Francis
>
http://www.guycox.com/optical.htm> ______________________________________________
> Guy Cox, MA, DPhil(Oxon), Honorary Associate,
> Australian Centre for Microscopy & Microanalysis,
> Madsen Building F09, University of Sydney, NSW 2006
>
> Phone +61 2 9351 3176 Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
>
http://www.guycox.net>
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:
[hidden email]]
> On Behalf Of John Oreopoulos
> Sent: Wednesday, 11 April 2012 10:29 PM
> To:
[hidden email]
> Subject: Re: Nyquist and Image size
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> *****
>
> Renato,
>
> Whether you have 256x256, 512x512 or 2048x2048, the "optimum" Nyquist
> sampling rate (ie: pixel dimensions) does not change since your objective
> lens did not change. The quoted pixel size at 2Kx2K you mentioned (22.5 nm
> x 22.5 nm) means you are oversampling the image (and not gaining anything).
> Your image may look smoother but it contains no more information than the
> 512x512 image with 90x90 nm pixel sizes. Presumably the scan speed is the
> same between 512x512 and 2Kx2K.
>
> You should decrease the galvometric mirror scan zoom setting to get back
> to an effective pixel size of 90x90 nm with 2Kx2K pixels in your image.
> Effectively, you will be imaging (and properly sampling) a larger field of
> view then. I'm not familiar with the Leica laser scanning confocals so I'm
> not sure if it will allow you to do this. On other systems, like the
> Olympus FV300 for example, you can set your image pixel dimensions
> (256x256, 512x512, etc.) and your scan zoom independently.
>
> Just out of curiosity, why image 2K x 2K when you can't easily display
> that on a standard computer screen or present it in a published paper
> without downsizing? I rarely departed from 512x512 in my laser scanning
> days, except when I wanted to see a larger field of view.
>
> Cheers,
>
>
> John Oreopoulos
> Research Assistant
> Spectral Applied Research
> Richmond Hill, Ontario
> Canada
> www.spectral.ca
>
>
> On 2012-04-11, at 7:22 AM, Renato Mortara wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > *****
> >
> > Dear all,
> >
> > Having attended the first Pawley course in Vancouver I feel highly
> > embarassed to ask this, but I would really appreciate a clarification:
> >
> > When estimating the highest zoom users should apply to their sample in
> order
> > to accommodate for the Nyquist theorem, I estimated the optimum pixel
> size
> > value by dividing the lateral resolution (eg: 0.2 microns) by 2.3 so that
> > the value is approxiametely 90 nm.
> >
> > The doubt: if the image size is increased from 512x512 (having adjusted
> the
> > zoom to the pixel size of 90nm) to 2Kx2K, the resulting pixel size
> > (displayed by the system - Leica) the pixel size decreases 4 fold, to
> 22.5
> > nm. Since the resolution obviously did not change but only the image
> size,
> > what happens to Nyquist and the optimum pixel size at 2Kx2K ?
> >
> > Many thanks !
> >
> > Renato
> >
> > Renato A. Mortara
> > Parasitology Division
> > UNIFESP - Escola Paulista de Medicina
> > Rua Botucatu, 862, 6th floor
> > São Paulo, SP
> > 04023-062
> > Brazil
> > Phone: 55 11 5579-8306
> > Fax: 55 11 5571-1095
> > email:
[hidden email]
> > home page: www.ecb.epm.br/~ramortara
>