Christophe,

                As I said in my original message, if you want to produce an image of your sensor, with the squares coloured in, that is indeed the case - no argument.  But that is not an image of your sample, it's an image of your detector.  I was (and am) writing about the digital image in general, as I thought I made clear, regardless of the detection technology.  Your CCD image does NOT tell you that intensity X exists across the whole of one detector element, and then suddenly intensity Y exists across the whole of the next detector element.  As you said yourself  "There is no information about where did the photons arrive on the square detector portion corresponding to a pixel".  So we are still dealing with samples and have to handle the problem of relating these samples to the real world.

                                                           Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier
Sent: Monday, 16 April 2012 10:52 PM
To: [hidden email]
Subject: Re: A pixel is not a little square

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Guy,

But in the CCD case the sampling is integrated over the whole pixel area of the sensor (assuming pixel coverage of 100% which is the case for EMCCDs for exemple). There is no information about where did the photons arrive on the square detector portion corresponding to a pixel. The value X you get as intensity for this pixel is the integrated density over the whole detector portion, so why would you take that value as a point-measurement corresponding to the center of this area ? In that CCD case, isn't the mage indeed an histogram with a bin that has (quite literally) a width equal to the pixel size, not a sample at mathematical points corresponding to the centers of the pixels ?

Christophe



On Mon, Apr 16, 2012 at 14:36, Guy Cox <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Mark,
>
>              You are continuing to confuse the samples with the
> representation of those samples.
>
> Let's imagine we have a series of data points:  100  90  80  70  60  
> 50  40
>
> We are mapping these on an image where each is separated by a defined
> distance.  So we need to fill in this distance.
>
> You are saying that the 'correct' representation is:
> 100  100  100  100  100   90   90   90   90   90   80   80   80   80   80
>   70   70   70   70    60   60   60   60   60   50   50   50   50   50   40
>
> I am saying that this is a wildly implausible and totally unjustified
> interpretation, and the best representation we can derive from the data is:
> 100  98   96   94   92   90   88   86   84   82   80   78   76   74   72
> 70   68   66   64   62   60   58   56   54   52   50   48   46   44   42
> 40
>
> EITHER way we are interpolating the sampled data - we have no option - so
> let's just get over this.   Your proposed representation includes detail
> that we could not possibly detect, mine does not.   Remember, these are
> SAMPLES.  Neither representation changes our recorded data.  End of
> story, IMHO.
>
> How did we get into this mess?  Why does everyone then 'do it wrong'?
>  Well, actually, everyone doesn't.  Scanning probe microscopes always
> remap
> - because by the time they appeared the computing power to do it was
> available.  When confocal microscopes first became widely available,
> in 1987, the data they produced completely overwhelmed available computing
> power (believe me, I was there and writing software).   So we got used to
> the 'quick and dirty' approach.  Consumer digital cameras do a sort of
> remap because the Bayer mosaic requires it, but modern sensors so far
> exceed the resolution of the camera optics that we never get to see
> any spurious frequencies anyway.  Computer games consoles always
> remap.  So do X-ray and EM tomography systems.
>
>                                                Guy (arrogant bastard)
>
>
>
> ----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]]
> On Behalf Of Mark Cannell
> Sent: Monday, 16 April 2012 7:37 PM
> To: [hidden email]
> Subject: Re: A pixel is not a little square
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Sorry Guy, I still think you don't see the point I'm trying to make.
> The camera actually says "The mean signal from x to x+dx is ..."
> (where dx is the sensor pixel size). It does NOT say the signal at x
> is 'K' and that is where I think the confusion lies. The camera output
> is a 2D 'histogram' and showing little boxes with the same intensity
> is (I say again) a perfectly accurate representation of the data (
> i.e. F(x) for x -> x + dx = K). With respect, it is not, as you say
> inaccurate -even if it is unaesthetic. If you fit a sinusoid you have
> just carried out a fitting exercise... That is not a "more accurate"
> presentation of the data despite what your Smith says (even if it may
> be a more accurate representation of the object which has been
> discretized). One should not loose sight of the fact that you have made some (possibly large) assumptions in the fitting process.
>
> Put mathematically, if you smooth out the displayed pixel edges you
> extend the actual sampling frequency (note how you are putting new
> unrecorded samples between recorded data values  -which is what
> drawing a line between points actually does) -you are adding
> information to the data that was NOT present in the RAW data. It may
> be that your additional information is correct and adds value (e.g.
> the band limit of the microscope is...) but one should not loose sight
> of distinction between the addition of data/information by the
> experimenter (which may or may not be wrong) and that reported by the
> instrument (the closest to truth the experimenter can get).
>
> At the risk of boring some readers on this list, let me emphasize my
> point
> : The camera actually says "The mean signal from x to x+dx is ..."
> (where dx is the sensor pixel size). It does NOT say the signal at x
> is 'K' . This can be portrayed as a square with constant color and I
> can think of no other truer portrayal of the measured data.  Hopefully
> dx is less than the resolution of the viewer at final display
> resolution but if it is not, then the only choice (IMHO) is between
> aesthetics (or some other goal) and truthfully displaying the recorded data -there is no middle ground.
>
> Cheers Mark
>
> PS My CD player can't output square waves because the detector etc.
> has a rather finite bandwidth... Even if it could, my ears are too
> many dB down at 44 kHz to sample it correctly and hear the artifacts
> introduced by digital sampling ... :-)
>
> On 16/04/2012, at 9:18 AM, Guy Cox wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > " There are _no_  'higher harmonics' present in the data, only in
> > ones
> 'artistic' interpretation for display purposes."  That is exactly what
> I said!
> >
> > I also never said that the data is a continuous function, I said it
> > is a
> series of discrete samples of a continuous function.  So when you
> choose to display it you have to do something.  Drawing little boxes
> is NOT 'doing nothing' and neither is it 'displaying the raw data'.  
> On the contrary -  it is corrupting the data with frequencies which
> shouldn't be there AND confusing the human eye (for which, presumably, we are doing the drawing).
>  The raw numbers are useful - indeed essential - for the computer but
> fundamentally cannot just 'be displayed' to the human eye as an image.  
> Our sampling rationale is based on sine-wave frequencies and
> therefore, as Alvy Ray Smith said, sinusoidal mapping is the truest
> (not the most aesthetic, though this is also true) way of displaying
> the data.  It doesn't add any spurious higher harmonics, it presents
> the data as accurately as our sampling permits.  Drawing little boxes
> may be easier, but it is just as much mapping the measured samples to
> a displayed image - the difference is that this method is both inaccurate and un-aesthetic.
> >
> > If your CD player spat out square waves to the speakers, you'd take
> > it
> back to the shop pretty promptly!
> >
> >
>                                                                   Guy
> >
> > -----Original Message-----
> > From: Confocal Microscopy List
> > [mailto:[hidden email]]
> On Behalf Of Mark Cannell
> > Sent: Monday, 16 April 2012 5:27 PM
> > To: [hidden email]
> > Subject: Re: A pixel is not a little square
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > I think I see the problem, the spurious frequencies arise from your
> thinking the _data_  is a continuous function and treating it as such
> (by "drawing a line ..."), but it is not, it  is discrete and can be
> faithfully represented by a _discrete_ Fourier transform (which folds
> at Fs/2). The hiighest frequency in the DFT is Fs, but we know we
> shouldn't look at that right?  There are _no_  'higher harmonics'
> present in the data, only in ones 'artistic' interpretation for display purposes.
> >
> > If it looks jagged, that is because in reality sampled data really is!
>  The problem really arises because you do not know how to fill in the
> space between data samples.  You can interpolate (or not). If you
> interpolate you are making a statement about the model underlying the
> data and have just carried out a fitting exercise. Fitting is NOT raw
> data presentation. If you just plot data values you make no assumption
> about what should join the data, no model has been fit to the data.
> Every scientist should know the difference between a histogram and a
> continuous distribution and not be fooled by the vertical lines at the
> histogram boundaries (which is what you show in a pixel image).
> >
> > The choice is yours, in one case you faithfully show unadulterated
> sampled data (the histogram looks less 'pretty' than a curve) or you
> fit a model and interpolate. The trouble with the latter is that the
> model is probably wrong and you hide the defects in the data (e.g.
> camera pixel
> size) from the keen eyed reviewer... Of course if the data points are
> really close together, the myopic reviewer can't see defects in you
> data
> :-) !  From  Guy's reasoning,  it would be impossible to represent any
> digitally sampled data because you are always pixelating a continuous
> function (all pictures get mad up of little squares -the printer dumps
> blobs of ink etc). So, where does the pixelation become acceptable?
> This is now aesthetic and has nothing to do with science or
> mathematics (those with perfect vision will always see discretization 'artifacts' more easily) .
> >
> > Cheers Mark
> >
> > On 16/04/2012, at 3:31 AM, Guy Cox wrote:
> >
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> *****
> >>
> >> OK, having slept on it, I now feel that just maybe I can explain
> >> what
> this is all about.  If only the list would let us include pictures it
> would be much easier!
> >>
> >> Let's assume we have a digital image, from any source, consisting
> >> of
> pixels with a spacing s.  The smallest spacing we can resolve in this
> image is 2s, and this will correspond, in frequency space, with a
> frequency f.  f represents the bandpass limit of this system,  no
> higher frequencies can be passed.  Now imagine we have a row of pixels
> containing the following
> values:
> >>
> >> 255  0  255  0  255  0  255  0  255
> >>
> >> If we represent these pixels by little squares, we'll have
> >> something
> like a chessboard.  Taking a line along this chessboard will give us a
> square wave.  Now this square wave cannot be represented within the
> bandpass limit of the system, defined by the frequency f.  To
> represent a square wave we need an infinite series of sine waves f + 3f + 5f +7f .....
>    To get even a crude approximation to a square wave we need f + 3f -
> that is a frequency three times higher than the image can contain.
> >>
> >> In other words, we've introduced a whole series of spurious
> >> frequencies
> into our image that not only were not there to start with, they could not
> possibly have been there.   Does this matter?  After all, we know they
> can't be real.  It does matter, because we are talking about a visual
> representation of our data - that's why we drew the little boxes in
> the first place.  Our eyes are very sensitive to edges* and the edges
> will take over if we let these frequencies come within the bandwidth of our eyes.
> We will find it very hard to actually see the finest detail in our
> picture (defined by 2s, remember) because if we enlarge it enough to
> see this easily we'll also get the edges created by these spurious
> frequencies.  In everyday terms, the pixellation takes over from the picture.
> >>
> >> Note that in all this discussion I have  not mentioned microscopes,
> cameras or anything - we are just talking about a digital image from
> any source.  It applies to confocal, widefield, and electron
> microscopes, telescopes, X-ray images and your holiday snaps.  Coming
> back to the microscopic world, if we oversample to the point where r,
> our minimum resolved distance, is substantially greater than 2s, we
> may not need to enlarge to the point where we see the spurious
> frequencies.  This is probably why some contributors to this
> discussion have advocated considerable levels of oversampling (though
> they probably didn't realise this, they just knew they got good
> pictures that way).  But oversampling in fluorescence can be very hard on our specimens.
> >>
> >> "But I'm using a CCD detector so my image is made up of little
> squares".  Yes, you can produce a 'coloured in' picture of your
> detector that way.  I'm assuming the image is actually what you want
> to see, though, not the detector.
> >>
> >> *Amusingly, the human eye does the same thing to emphasize edges as
> computer image processing does - it makes the dark side of the edge
> darker than it is and the light side lighter.
> >>
> >>
>                                                          Guy
> >>
> >> PS.  This has doubtless confirmed my reputation among some people
> >> as an
> arrogant bastard.  They are probably right, but at least I'm an
> arrogant bastard who tries to help.  It's taken me two hours to write this.
>