> There are layers in new lenses that can not stand organic solvents as
> Acetone, be careful with that, you might soak a color filter, and end up
> with transparent glass, better ask them about the model you purchased. Best
> regards Gabriel OH
> > Date: Tue, 26 Jun 2012 14:24:33 +0000
>> From:
[hidden email]
>> Subject: Re: Removable glue for optics
>> To:
[hidden email]
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>>
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> *****
>>
>> He everybody,
>>
>> Thanks a lot for all the answers - very useful!
>>
>> I am a bit reluctant to rely on solvents like acetone to remove the glue,
>> though. Does anyone know if the surface coating of standard chroma
>> dichroics would survive acetone treatment?
>>
>> I think I will go with silicone in the corners and edges of the dichroic /
>> aluminum interface for now. I just hope it's not getting sucked under the
>> glass...
>>
>> Just to sum up the responses:
>>
>> Collodian (wow - that stuff is ancient!)
>> Crazy Glue
>> Duco cement
>> silicone (e.g 732 RTV)
>> rubber cement
>> two sided scotch tape
>> paraffin
>> dental wax
>> postit note
>> gelatin
>> PVA blobs
>> Hot glue
>> Nail varnish
>>
>> Best,
>> Tobias
>>
>> --
>> Tobias Rose, PhD
>> Max-Planck-Institute of Neurobiology
>> Department of Cellular and Systems Neurobiology
>> Am Klopferspitz 18
>> D-82152 Martinsried
>> Phone: +49 89 8578 - 3684
>> Fax: +49 89 8578 - 2481
>>
[hidden email]
>>
http://www.neuro.mpg.de/37634/rose>>
>>
>> > -----Original Message-----
>> > From: Confocal Microscopy List
>> > [mailto:
[hidden email]] On Behalf Of Dan Focht
>> > Sent: Tuesday, June 26, 2012 8:26 AM
>> > To:
[hidden email]
>> > Subject: Re: Removable glue for optics
>> >
>> > *****
>> > To join, leave or search the confocal microscopy listserv, go to:
>> >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> > *****
>> >
>> > Tobias
>> >
>> >
>> > This is old but it works for temporary bonds.
>> > Collodian.
>> >
>> > Dries fast, light to medium hold, dissolves easily and most
>> > importantly,
>> > cheap!
>> > Caution, flammable when in liquid form.
>> > You might find it in the OR of a hospital.
>> >
>> >
>> > Dan
>> >
>> >
>> > On Jun 25, 2012, at 11:38 AM, Tobias Rose wrote:
>> >
>> > > *****
>> > > To join, leave or search the confocal microscopy listserv, go to:
>> > >
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy>> > > *****
>> > >
>> > > Dear all,
>> > >
>> > > I need to glue some glass optics onto anodized aluminum surfaces
>> > > (edges of coated dichroic mirror surface and of a full reflective
>> > > silver mirror). The problem is that I'd like to be able to remove the
>> > > mirrors at some point without completely ruining their reflective
>> > > surfaces and leaving too much residue on the glass and aluminum.
>> > > The glued parts are not in the optical path so the glue does not have
>> > > to be transparent. Also the glue can be quite weak, just enough so
>> > > that the mirrors don't fall off.
>> > >
>> > > Can anyone suggest a nice glue for that?
>> > >
>> > > Best,
>> > > Tobias
>> >
>> > Dan Focht
>> >
[hidden email]
>> > Bioptechs, Inc.
>> > 3560 Beck Rd.
>> > Butler, PA 16002
>> > www.bioptechs.com
>> > Direct 724-282-7145
>> > Fax 724-282-0745
>> > Toll Free 877 LIVE-CELL - (548-3235)
>