Ramana,
We addressed this issue on a stereo dissecting microscope by detecting the
GFP using a YFP filter set. If you're using filters for your imaging, you
can try this route.

Good luck,
~Gregg
*Gregg Sobocinski*
Microscope Imaging Specialist
University of Michigan, MCDB Dept.
Ann Arbor, Michigan, USA

On Wed, Aug 1, 2012 at 4:44 PM, Julio Vazquez <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Separating the two proteins should not be too difficult if you have the
> right filter sets, or if you have a spectral microscope system. If you
> check the spectra (e.g. invitrogen spectra viewer which you can Google),
> you will see that GFP emission starts at ~ 500 nm, so a narrow CFP filter
> should give litttle or no GFP signal. Conversely, if you excite GFP at
> around 488 nm, there should be little or no CFP excitation.  Your
> institution probably has a Microscopy Core facility, so the best is to talk
> to them... I'm sure they can point you in the right direction.
>
> --
> Julio Vazquez
> Fred Hutchinson Cancer Research Center
> Seattle, WA 98109-1024
>
>
> http://www.fhcrc.org
>
>
> On Aug 1, 2012, at 12:55 PM, Ramana Sidhaye wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Hi,
> >  One of my collaborators has a mouse endogenously expressing 2 proteins
> (both are expressed in the same cells), one coupled to CFP and the other to
> GFP. I am curious about the spectrum- how does one best evaluate both
> proteins in the same sample without concerns about cross-reactivity.
> >
> > Thanks,
> > Ramana SIdhaye
>



-