Posted by Kate Luby-Phelps on
URL: http://confocal-microscopy-list.275.s1.nabble.com/air-table-floatation-compressor-vs-house-air-tp7578794p7578801.html

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

We have used house air with a water trap for years on our LSM 510 with no problem.
However, if you plan to use house air, check first to be sure the pressure is high enough (80
psi) and put a regulator on it so you can tune the table.

Kate L-P